ICP-MS 8900 1233 error Rf has switched 48V shut off. Plasma cannot be ignited. Can someone tell me why and how to solve? Thank you!
Find maintenance, best practice, and troubleshooting articles.
ICP-MS 8900 1233 error Rf has switched 48V shut off. Plasma cannot be ignited. Can someone tell me why and how to solve? Thank you!
Hello, Error 1233 refers to some pre-requirement interlocks not meeting as a result During the ignition of plasma the RF generator shuts down the +48 V power supply. Can you please re-verify if there was no drop in argon pressure, the plasma door, and the analyzer door were properly fixed, and the cover interlock was OK?
Thank you for reply! Argon pressure dropped a lot and quickly during that short period of time.
Thank you for reply! Argon pressure dropped a lot and quickly during that short period of time.