ICP5800

I need to set up a new method for aqueous analysis, and would like to calibrate the instrument for 18 analytes with a multielement standard; however, I do have a lot of interference with my wavelength selections using ICP Expert. Can anyone recommend a way to set up the method?. Thanks for the help. 

Parents
  • If you have the latest version of ICP Expert (7.6) then you could run the standard and/or a sample with IntelliQuant screening.  After you save the IntelliQuant Screening worksheet, you could create a new worksheet using the recommend lines from the IntelliQuant screening worksheet. IntelliQuant recommends the best wavelengths based on freedom from spectral interferences.

Reply
  • If you have the latest version of ICP Expert (7.6) then you could run the standard and/or a sample with IntelliQuant screening.  After you save the IntelliQuant Screening worksheet, you could create a new worksheet using the recommend lines from the IntelliQuant screening worksheet. IntelliQuant recommends the best wavelengths based on freedom from spectral interferences.