Measuring Cr in high ferrous material

Hi
does anyone have an explanation for the fact, that I have a lower result for Cr (approx. 30%) with material with a high iron content, e.g. Millscale, with the measuring mode 'NoGas'? Measured with He mode, the results are correct. Measured with ICP-MS Agilent 7800, digestion in HNO3, HBF4 and water.
Thank’s and regards

Parents Reply Children
  • Hello

    I am not an applications expert,  but if I had to guess it is because ArC would be present through your calibration - from blank to all standards. ArC present in no gas mode would artificially increases counts in blank and standards, thus potentially giving you a lower check standard concentration

    I found this great Agilent applications note further explain Cr52 interference. Hope this answers your question