Is there an interferent equation for Iron No Gas mode for the Masshunter software?
Find maintenance, best practice, and troubleshooting articles.
Is there an interferent equation for Iron No Gas mode for the Masshunter software?
No there is not one for that because the main interference is Argon40.
Most customers use Fe57 if they have to use NoGas
No there is not one for that because the main interference is Argon40.
Most customers use Fe57 if they have to use NoGas