Can I use Gold Chloride when avoiding mercury memory effects when doing ICP-MS analysis?
Find maintenance, best practice, and troubleshooting articles.
Can I use Gold Chloride when avoiding mercury memory effects when doing ICP-MS analysis?
What is the best concentration to use? Do I have to add it on wash vails not the original samples and calibration samples?
we
use 0.5% hcl
i
use in all my rinses
as
for gold concentration i’m not exactly sure as that can vary
So do I use 0.5% HCL as a rinse solution? So is it better to use this solution after every 5 samples ran?
Yes you can use 0.5% HCl as the rinse solution.
Yes you can use 0.5% HCl as the rinse solution.