MP-AES auto - background correction feature

Hi everyone in the community, 
While I was performing an analysis with the MP-AES I noticed that the linearity of my calibration curve for zinc was quite high (R=0.997) when I used the auto background correction feature in the MP Expert software compared to the non-linearity (R=0.53) I observed when background correction was turned off so that I could look at the raw data.

 

My question is how does the auto background correction work to reduce interference and noise to isolate the signal so adequately? 

The description in the software states that it calculates the concentration using the total intensity but is there any more to this fact? I've searched online but also couldn't find an answer.

 

I would greatly appreciate some feedback! 


Thanks all,


Shane Grant

Parents
  • Hi Shane,

     

    How is the non-linearity line performed, without the background correction with off-peak?

     

    The background correction of the MP Expert software is dynamic and based on the Blank and Standards graphic. It's very important that the Blank is the same matrix as the samples and that the lowest standard has a quite good defined peak shape.

    With the black graphic compare to the difference in shape of the standards the software made a background profil. This profile could be seen if the "Black Subtraction" is set off.

     

    The linearity for the MP-AES should be very high and quite long if you have wel defined peak shapes and if you don't have spectral or physical interference. 

     

    is this answer you question?

Reply
  • Hi Shane,

     

    How is the non-linearity line performed, without the background correction with off-peak?

     

    The background correction of the MP Expert software is dynamic and based on the Blank and Standards graphic. It's very important that the Blank is the same matrix as the samples and that the lowest standard has a quite good defined peak shape.

    With the black graphic compare to the difference in shape of the standards the software made a background profil. This profile could be seen if the "Black Subtraction" is set off.

     

    The linearity for the MP-AES should be very high and quite long if you have wel defined peak shapes and if you don't have spectral or physical interference. 

     

    is this answer you question?