Hello payninater ,
It is not possible to enter responses manually in the Method Editor.
As far as I know there should be no issue with setting your sample type and level in the batch table, creating a new method from acquired data, creating levels…
Hello HowngChan ,
For a dual needle multisampler the purge timer does not start until after the method with a different gradient or solvent system is loaded. By default, the worklist will have a 10 minute Wait Time for Ready. This can be set to a maximum…
Thanks for your reply. The syringe O rings are quite old so I will change them. I was able to overcome the issue by using a pulsed split mode on the inlet. Your other points are very helpful info and give me things to investigate.
Browse the help documentation by using the navigation menu or the links below.
Agilent Community Best Practices
Forum Best Practices
How to Post a Question
How to Reply to a Question
How to Attach Files
How to Configure Notifications…
Use the navigation menu to view our help articles.
Customize Your Email & Live Alert Notifications
Subscribe to Areas of the Agilent Community
Turn on or off Email Notifications
Which Electron Ionization Source Part 1 had questions. Which Electron Ionization Source Part 2 had some answers. Now, let’s talk about each source and tuning.
Stainless Steel/Inert vs. Extractor, Atune vs. Etune - The colored lines in the diagrams …
Hello MLGonzalez ,
In place upgrading for quant was only supported starting at Version 12.0. If you have any versions prior to that you must first uninstall quant before attempting to install a different version. If this was Quant 11.0 for TOF/QTOF…
This Information Applies To: HPLC Advisor app
Issue
The Agilent InfinityLab HPLC Advisor app is a cost-free mobile app for both Android and iOS.
It offers tools to save time with HPLC troubleshooting, method development, and more. Once installed...
This Information Applies To: Agilent 5977A/B GC/MSD with MassHunter GC/MS Acquisition Software
Issue:
This article describes the procedure to safely pump down (start) 5977A/B GC/MSD with Electron Ionization (EI)
Steps To follow:
Before you...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article...
This Information Applies To: Agilent 5975 and 5977 series GC/MSD
Issue:
If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant.
Inspect the calibration vial on the back of the flow controller...
This Information Applies To: Agilent 5975 and 5977 Series GC/MSD
Issue:
For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation...
This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD
Issue
Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they...
This Information Applies To: Agilent GC/MSD systems
Issue:
After a filament fails, there will be a system error message: "No emission current". Switching to the other filament (except for the CI ion source, which has just one filament), and then...
This Information Applies To: Agilent 7693A automatic liquid sampler tray
Issue:
The orange Align Mode light comes on . T he sample tray of the Agilent automatic liquid sampler 7693A will not operate again until the alignment procedure has successfully...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide...
This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
Two turrets are provided with the 7693A injector, the 16-sample standalone turret, and the 3-sample transfer turret used with the 150-sample tray (See Figure 1 ). The turret...