I guess checking the settings will be the first thing. On the GC side, since RT did not shift too much, I think oven temp would be of a less issue. But flow and pressure might act more in this case. Based on info I learnt from here, both pressure and…
Hello AJohnson13 ,
Lab Advisor can be currently downloaded from this page.
Lab Advisor Software Download | Agilent
This may change over time. You can search the web for 'agilent lab advisor download' to find a link to the current download location…
Hello,
So, you should be able to add another column to your report table and change the value to = Compound_Concentration * Sample_Amount / 100 to remove the sample amount and percentage. If you need to remove a multiplier or dilution you will need…
Hello,
LIMS ID is not available as a token for sample name, datafile name, or in CDS DA for report output. You can see the options in the screenshot below. It can be a value in the CSV file but not used in the name. The only other option would be…
Our G4220A (1290 Bin Pump) is setup to work with the seal wash currently. It operates with our instrument with no issue. If I go into the instrument configuration and check the box to enable the pump seal wash (The pump is installed), it will lose connection…
Hi RDP0314,
One thing to check is the injection settings. This behavior is common when the GC is expecting a series of manual injections. To check, click on the Instrument menu shown on your screen capture above and 'Select Injection Source' or similar…
This Information Applies To: 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
Issue:
The Peristaltic Pump needs to be replaced if it has stopped working due to worn tubing, if it is leaking, or as part of annual maintenance of the...
This Information Applies To: Agilent Autosamplers G1313A, G1329A/B, G4226A, G1367A/B/C/D/E, G7129A/B/C.
Issue:
If during initialization or during a sequence, the "Metering Home Failed" error occurs, this article will guide through the procedure...
This Information Applies To: 1100, 1200 and 1260 Infinity I and 1260 Infinity II LC Pumps (Excluding preparative pumps with electromagnetic purge valve (EMPV): G1361A. Not applicable to any of the 1290 Infinity II modules with automatic purge valve: G4220X...
This Information Applies To: Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
Issue:
Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules.
Both the visible and UV lamps are consumable...
This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pumps.
Issue:
Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module.
Steps...
This Information Applies To: All Agilent 1260 Infinity Binary, Quaternary, or Isocratic Pumps without the Seal Wash Option installed.
Issue:
Worn pump seals can shed small particles that can cause blockages and pressure issues downstream of the...
This Information Applies To: Agilent G7167x Multisampler modules
Issue:
Required when:
The needle seat is visibly damaged, blocked, or is leaking.
Leaks or blockages in the Vialsampler are observed, especially between the needle and needle...
This Information Applies To: Agilent 1260 Isocratic (G1310B) Quaternary (G1311B) Pumps.
Issue
Required when:
The inlet valve is defective.
The pressure or leak test in Lab Advisor suggests the inlet valve may require replacement.
A large...
This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules
Issue:
Required when:
The outlet valve is defective or leaking.
The pressure or leak test in Lab Advisor suggests the outlet valve may require replacement...
This Information Applies To: Agilent 1260 Isocratic (G1310B) quaternary (G1311B) pump modules
Issue:
The seal wear in procedure is important as it acts to "bed-in" the seals to ensure a leak free and pressure-tight pump head.
Required when:...