Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Injectors and Sample Thermostats
  • Product Name :  1100 Series LC System 1200 Series LC System 1260 Infinity LC System 1290 Infinity LC System
  • Task :  Maintenance
Related Articles
  • Removing and Reinstalling the Needle Seat in Agilent G4226X/G1367E High Performance (HiP) Autosamplers

    This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosampler Issue Replacing the Needle Seat of the Autosampler is required when: The needle seat is visibly damaged. Leaks or blockages in the Autosampler are observed from the needle seat. Replacement the needle seat is performed as part of yearly maintenance. Tool Required Wrench, open end 1/4 - 5/16 inch (p/n 8710-0510 ) Parts Required For an Agilent 1290 Infinity (G4226A...
  • How to Remove and Reinstall the Needle in Agilent G4226X/G1367E High Performance (HiP) Autosamplers

    This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosamplers Issue Replacing the Needle Assembly of the Autosampler is required when: The needle is visibly damaged. Leaks or blockages in the Autosampler are observed from the needle. The limit for the needle interaction EMF counter is exceeded. Replace the needle as part of yearly maintenance. Tool Required Wrench, open 1/4 - 5/16 inch (p/n 8710-0510 ) Parts Required...
  • Troubleshooting "Metering Home Failed" Error in Agilent Autosamplers

    This Information Applies To: Agilent Autosamplers G1313A, G1329A/B, G4226A, G1367A/B/C/D/E, G7129A/B/C. Issue: If during initialization or during a sequence, the "Metering Home Failed" error occurs, this article will guide through the procedure of identifying the cause. Sometimes, this error may also appear as "Cannot Find Metering Sensor" . Steps to follow: Often the reason is a blockage in the injection system. Remove the capillaries into and out of the metering device ( Figure 1...
  • How to Remove and Reinstall the Loop Capillary in G4226A/G1367E High Performance (HiP) Autosamplers

    This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosampler Issue Replacing the Loop Capillary of the Autosampler is required when it is broken or leaking. Items Required Wrench open 1/4 - 5/16 inch (p/n 8710-0510 ) For an Agilent 1290 Infinity (G4226A) HiP Autosampler, one of the following: Loop cartridge 20 µL, Red Capillary (p/n G4226-60310 ) Loop cartridge 40 µL, Green Capillary (p/n 5067-4703 ) Loop cartridge 100 µL...
  • Comparing Multidraw Injection and 900 µL Analytical Head for High-Volume Injections

    This Information Applies To: Agilent Analytical LC Autosamplers Issue To increase the maximum injection volume of an autosampler, there are two ways to do it. While both will provide larger injection volumes, there are advantages and disadvantages of each approach. This article will provide some insights of choosing either Multi-Draw or 900 µL analytical head. The 900 µL option will change the analytical head as well as adding a larger injection loop to allow 900 µL injections in a single...

How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)

Answer

This Information Applies To: 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)


Issue:

The Peristaltic Pump needs to be replaced if it has stopped working due to worn tubing, if it is leaking, or as part of annual maintenance of the autosampler.


Steps to follow:

 Warning 
Warning: Make sure that appropriate personal protective equipment to manipulate the chemicals used in this process. Nitrile gloves (or equivalent) and safety glasses are recommended.

 

 Caution 
Caution: Please note the directional arrow on the peristaltic pump showing the direction of pump flow when the peristaltic pump is turned on.


1. To remove the peristaltic pump, pinch the clips at the top and bottom of the peristaltic pump and pull the pump straight off the motor rod.

 

2. Remove the two tubes connected to the peristaltic pump by grasping and pulling on the tubing to pull it out past the spring on the peristaltic pump tubing. One tube goes to the rinse port, and the other goes to the solvent bottle.

 

3. To replace the peristaltic pump, first install the tubing by firmly pushing into the peristaltic pump connector. The tubing from the solvent bottle goes to the bottom connector of the peristaltic pump, and the tubing going to the rinse port, connects to the top connector of the peristaltic pump.

 

4. Line up the hole on the peristaltic pump with the motor rod. Firmly push the new peristaltic pump straight onto the motor rod until the clips at the top and the bottom of the peristaltic pump fully engage with the slits in the front wall of the autosampler. You should feel it snap firmly into place.

5. Using either your CDS or the Instrument Control section of Lab advisor, turn on the peristaltic pump. If there are no leaks, and the flow runs from the solvent rinse bottle to the rinse port of the autosampler, the peristaltic pump is correctly connected.

 

 Tip 
To see a list of consumables and supplies to maintain your Agilent Autosampler module:
HPLC Autosampler and Fraction Collector Supplies accessories page on Agilent.com
 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help