Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
This Information Applies To: Agilent CI Source
Issue:
During normal operation of the GC/MS, the CI ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filament and...
This Information Applies To: Agilent OpenLab CDS
Issue
This article will explain how to add a sequence summary report line to a sequence, in Agilent OpenLab CDS, as well as how to configure it so it matches your reporting needs.
Background ...
This Information Applies To: Agilent OpenLab CDS, OpenLab CDS ChemStation, OpenLab CDS EZChrom.
Issue
Process for creating a desktop shortcut for your instruments, with OpenLab software, to enable quick access to your acquisition window.
Background...
This Information Applies To: Agilent LC Systems
Issue
Agilent modules will go into an error status (red status light) when an error condition specific to that module occurs. For example, metering home failed will be specific for the autosampler...
This Information Applies To: Agilent Analytical LC Autosamplers
Issue
To increase the maximum injection volume of an autosampler, there are two ways to do it. While both will provide larger injection volumes, there are advantages and disadvantages...
How to Recover a SubscribeNet Password
Multiple Results
This Information Applies To: Agilent Software products with SubscribeNet license.
Issue
Resolutions for password recovery use if you have: lost your password, been locked out of the account, or you don't know if you have an account with SubscribeNet...
This Information Applies To: Agilent OpenLab CDS, OpenLab CDS EZChrom
Issue
Steps to create a Data Analysis shortcut for one or more of your Projects defined in the OpenLab Control Panel .
Steps to Follow
Open the Control Panel of your...
This Information Applies To: Agilent 6100, 6200, 6400 and 6500 LC/MS Systems
Issue
Error on the Mainboard reporting “Leak Detected”.
Background
The fault “Mainboard #1: Leak Detected” is reported in the Data Acquisition Software Logs ( Figure...
This Information Applies To: Agilent LC/MS Systems
Issue
The MS solvent selection valve (SSV) rotor seal is replaced annually as part of the Preventive Maintenance. It can also be replaced as part of troubleshooting a leak, a flow blockage, or when...
This Information Applies To: All Agilent HPLC modules.
Issue:
The firmware is stored on a HPLC module's mainboard. It's a set of parameters, low-level commands, and functions stored on the internal memory of a HPLC module and required for its operation...