Did you miss April's GC Live Session Recording and Discussion Thread? Check it out here: ( +) Diagnostics From Your Desk, Troubleshooting From Your Kitchen Table - Follow-up Discussion…
Hello cjwwfd ,
B.06.00 is out of support at this point, but I can offer some general suggestions for current revisions and hopefully you can apply them to your version. Version 12.0 is the current release. If possible, I would investigate your options…
I'm experimenting the same error. Logbook displays <MSQQQ_1: QqQ fault detected: 1.23 vacuum system is not ready. This is typical during a pump down cycle or when the high vacuum pump is turned off during a vent cycle or fault condition>. Does anyone…
Hello DET ,
Sorry for the late response on this.
What sort of spiking do you see usually? What is the output of the FID, in pA?
Are you using any UPS systems, or are the GCs plugged directly into the wall?
Hi, I think it would be a good idea to upload your startup tune report in order to help you better.
Also I think you could take advantage of reading this post: https://community.agilent.com/technical/atomic-spec/f/forum/470/icp-ms-manual-tuning
Hello,
So, file based workstations store the full path to the methods in the sequences. If there is any change between the systems in the project path this is the expected behavior if you copy a sequence. If you open the sequence, browse out and find…
Hello guys,
I recently discovered, that it is possible to change your font properties in the injection list via shortcut (ctrl + b for "bold" and so on..).
Therefore I wanted to ask you, if there is some code for changing font types for CCF outputs…
Hi Mai,
Thank you for reaching out to Agilent technical support.
Please check:
if the data file you are trying to open was generated using TapeStation security module as that can be one reason for the file not to open and error out.
If anyone…
Thank you for confirming what I had found. I figured that there wasn't a way around the time reference/ISTD link. Mostly, I am surprised that MassHunter is limited in a way that Chemstations/OpenLab are not.
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items...
This Information Applies To: Agilent GC Systems with flame ionization detector (FID)
Issue
Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items required...
This Information Applies To: Micro Electron Capture Detector (uECD)
Issue
Baking the micro electron capture detector (µECD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Warning ...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue
This article will describe how an operator can configure the makeup gas on both the instrument keypad and the software driver (Method Editor).
Background...
This Information Applies To: Agilent HPLC systems
Issue
The presence of ghost peaks can be a symptom of a system contamination issue, caused by accumulation of sample and/or foreign substances in the flow path. It is critical for high sensitivity...