We have not tested nor validated the use of Oligo Pro II data on ProSize so we cannot comment on why it would work for your colleagues. As for why only a single row is showing, please collect the zip file for your data as outlined in my previous message…
At 310° C ion source temperature, the 502 will always be lower than typical. Also - please reduce the quad temperature down to 150° or maybe as high as 165° C... High quad temperature reduces the lifetime of the quad and only very few applications need…
Hi Andreja,
For calculating area % of two different peaks you must set one baseline left and one baseline right from the peaks. Then you have to enable "Information Pane" when clicking "View" in the menu bar and the results are shown as given in the…
Hi,
Yes it is right, please orientate on the last peak before the impurities start.
The exact baseline settings can be seen when enabling the peak information section below the analysis view (highlighted in green):
Manual integration limits…
We were able to configure the plate, we are good for now.
The 1290 Infinity II we have is coupled to Sciex 6500 MS and is controlled by Analyst Software using the Analyst Device Driver.
Hello,
No, you cannot use classic report styles *.frp from Chemstation in the CDS 2.x software. You can use any IR template built in Chemstation in CDS 2.x by copying the .rdl files over to the new project folders.
Marty
The software (except for MicroLab Pharma) comes with a 90-day trial period before the software needs to be registered. If you get in contact with our sales team, they should be able to get you a trial copy of the software. Find the appropriate contact…
Hello, Is it possible to get the Intelligent rinse working if the rinses occur after the analysis. Let's say I have 2min analysis time and I changed the Probe rinse of samples and stds to 120s and the Probe rinse 1-3 after the initial rinse. Then the…
This Information Applies To: Agilent OpenLab CDS
Issue
Identifying the baseline codes and baseline code descriptions for OpenLab CDS
Background
OpenLab CDS Data Analysis includes the same integration algorithms as available in both OpenLAB...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to filter a signal in an Intelligent Report.
Requirements
To be able to carry out the procedure as described, the privilege of Data Processing...
This Information Applies To: Agilent OpenLab CDS
Issue
This article outlines different ways to quantify the amount of compound in an unknown sample.
Background
Quantitation uses peak area or height to determine the amount of a compound in...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle seat of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosamplers
Issue
Replacing the Needle Assembly of the Autosampler is required when:
The needle is visibly damaged.
Leaks or blockages in the...
This Information Applies To: Agilent 6000 Series LC/MS Systems equipped with MS40+ rotary vane pumps
Issue
Replacing the MS-40+ (rough pump) fluid periodically is a key maintenance procedure to improve vacuum performance and minimize the possibilities...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here
Issue...
This Information Applies To: Agilent 6000 series LC/MS instruments
Issue:
Cleaning the ion transfer capillary as part of regular maintenance or as part of troubleshooting when a loss of sensitivity in the instrument occurs.
Steps to follow ...
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Efficient electrospray ionization (ESI) in your LC/MS ion source requires a properly adjusted nebulizer needle in good condition. The nebulizer needle is a consumable item in...