Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
Hello Samehalsharawy ,
Can you please elaborate on the statement we uninstalled the cylinder expecting the flow to come down to zero but it didn't. It stayed at 50?
If no physical tank is connected to the GC, the GC gas connections are capped…
This Information Applies To: Agilent OpenLab CDS (version 2.x)
Contents:
In the following procedure, you will find steps to create a multipoint ESTD calibration method with OpenLab CDS 2.X.
Steps to follow:
1. Load the Data
a. From Data...
This Information Applies To: Agilent MassHunter Quantitative Analysis Software
Issue
Chromatographic peak integration is a critical part of accurate and reproducible quantitative data analysis. The integrator type and the integration parameters...
This Information Applies To: Agilent 6200/6400/6500 Series LC/MS Systems
Issue
Instructions on how to fill the calibrant bottle on 6200/6400/6500 Series LC/MS Instruments when calibrant levels are running low.
Background
The ability to autotune...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal.
Issue:
Installing the Ion Transfer Capillary after cleaning or replacement. For instructions for installing an ion capillary...
This Information Applies To: AriaMx Real-time PCR System
Issue
The saturation level is unique for each optical module after the offset of calibration factor and background calibration. The FAM, ROX, HEX, CY3 & CY5 optical modules will saturate at...
This Information Applies To: Agilent LC Systems
Issue
While it is possible to combine modules from different HPLC series such as 1100, 1200, 1260 or 1290 series to create a functioning LC system, there are some important topics to consider.
...
Fixes for the "RPC Server Unavailable" Error
Multiple Results
This Information Applies To: Agilent OpenLab CDS
Issue
When launching the instrument Acquisition screen/instrument session, there is an "RPC Server Unavailable" error.
Resolution
Accessing the Services App (Windows)
There are two ways to...
This Information Applies To: Agilent LC Pumps
Issue
If a user suspects that the gradient composition isn’t accurate and causes poor chromatography such as retention time shifts, late, early, no peaks or merging peaks. This article provides a quick...
This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosampler
Issue
Replacing the Needle Seat of the Autosampler is required when:
The needle seat is visibly damaged.
Leaks or blockages in the...
This Information Applies To: Agilent OpenLab CDS
Issue
This article discusses different types of bracketing modes and how to set up a bracketed calibration in Data Analysis for OpenLab CDS 2.x
Background
Occasionally, the instrument conditions...