Hello Kirill_M ,
The averages for response or concentration are not available in the batch table. There have been enhancement requests made to add information like this to the batch table, but there is currently no timeline for when this will occur…
Support and educational resources for the PlateLoc Thermal Microplate sealer and PlateLoc Microplate Heat Seals .
Support Documents
PlateLoc Thermal Microplate Sealer Site Preparation Checklist
PlateLoc Thermal Microplate Sealer Installation Checklist…
For the GC-6890 FID detector issues you're experiencing, here's a simplified approach to troubleshoot the errors:
Gas Supplies:
Make sure all gas supplies (utility, makeup, hydrogen) are on.
Check that the pressure…
I see the clear difference between the split mode vs the splitless that now has the valid parameter. and the initial nitrogen elution seems a lot less problematic. thank you for your instructions!
Hi rguzmans1
To prepare a litre of this mobile phase, weigh 0.315g of ammonium formate and dissolve in 800mL of water, add 100µL formic acid and mix well, then make up to total volume of 1L.
Probably best to filter this after preparation through 0…
This Information Applies To: Agilent GC Systems
Issue
Excessive baseline noise can be introduced at different points including sample preparation, sample introduction, separation, and the detector. The excessive background can cause reduced signal...
This Information Applies To: Agilent GC Systems
Issue
Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability...
This Information Applies To: Agilent GC Systems
Issue
Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
This Information Applies To: Agilent GC Systems
Issue
A shift in retention time can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
A change in peak size can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder.
Other chromatographic problems are identified in Basic...
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...