We attempted the same testing on another OL2 instrument and the sensitivity issue was not present. It was unique to a single system. Agilent was very helpful getting us another system to test on. It seemed to be related to the tune/threshold/filtering…
Hi,
1. This is the "Easy Fit Lense Assy" - Since two years, we (Agilent) has decided not to clean the lenses, but to replace them completely. So if you want to disassemble these, you have to break the plastic cap, and use a screw set from an older regular…
Hello,
You would need to look into the other logs to see the issue in more detail. The details logs are in a Microsoft log format and required a viewer download. I would suggest opening a ticket with your local Agilent support group. It does look…
Hello,
I cannot give you a direct link for the software. You would need to use the Agilent subscribenet to download any software where you have an entitlement. This would be any software you have registered with the site for licensing in the past…
Can you prepare the alkane standard in a different alkane, maybe heptane or octane? Boiling point of n-pentane is fairly close to room temperature (36 C) so you may be getting some evaporation of the lighter components in the sample. If there's enough…
Very useful video for basic understanding of the valve functionality. can you post some more videos on GREEN HOUSE GAS ANALYZER or TOGA System in this impressive and easiest way only.
This Information Applies To: Agilent GC Systems
Issue
Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
This Information Applies To: Agilent GC Systems
Issue
A shift in retention time can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
A change in peak size can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder.
Other chromatographic problems are identified in Basic...
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...