In spectroscopy, the quality of spectral data depends greatly on the incident beam that passes through the sample being analyzed. For samples analyzed at large angles of incidence (AOI), such as bandpass and edge filters (e.g., beam splitters), maintaining…
In my case, the problem was due to an account that could not log in to the company-specific drive where all measurement results are saved. If you have the same problem, try another account and it should work.
I will list a few common errors with this. If it isn't one of the common errors, you may need to get tech support involved to fix it because the software may not be working properly.
Is solvent flowing properly? When you tell one line to be used at…
Dear, Sr
When I want to go to execute any program in WinUV software, it gives a message "Error code 9:1 CSM error,CSM error"
It seems a comunnication error, the instrument is connected via GPIB-USB-HS.
Anyone knows how to do?
I assume it is packed tight with insulation? Otherwise I'd just call Agilent for any ideas, maybe a board is bad or something and you need a service call.
This Information Applies To: Agilent GC Systems
Issue
Excessive baseline noise can be introduced at different points including sample preparation, sample introduction, separation, and the detector. The excessive background can cause reduced signal...
This Information Applies To: Agilent GC Systems
Issue
Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability...
This Information Applies To: Agilent GC Systems
Issue
Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
This Information Applies To: Agilent GC Systems
Issue
A shift in retention time can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
A change in peak size can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder.
Other chromatographic problems are identified in Basic...
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...