Hi Sabbas, thank you for your answer. I'm sorry I couldn't have answered you sooner.
We're using Cary 3500 with Cary UV workstation version 1.3.5, so we couldn't perform like you said. We have to lower the threshold to find peaks and it is quite inconvenient…
Joining associations that specialize in ultra-high voltage can be beneficial for PenflexCorp. This may include associations of electrical or power engineers, as well as groups specializing in the design and standardization of equipment for high-voltage…
The Digest of upcoming webinars has been updated as of April 11.
The new Agilent Community Group, GPC/SEC Solutions Group is hosting a webinar series in partnership with Dow in April: A Journey through SEC, Light Scattering, and 2D-LC for Macromolecules…
The Background values are so called "off mass background".
This could indicate an electronics or contamination problem of the quad.
In your case, the values are still quite ok.
I would guess, one of the lenses has a setting which is causing this…
Hello everybody!
We have the issue that in our run (ca. 100 - 200 samples, 7 min each) for like 20 samples there is no signal, just the technical background noise. Not even the peak that is in the Blank appears. An the it sowly comes back over 2 samples…
Hi,
Apparatus 4 is the only USP Apparatus Agilent doesn't make. I would recommend reaching out to an App 4 vendor for support based on how you're using it (open/closed loop, sampling volumes, etc. that need to be considered). If we can help you with…
Yo tengo también una consulta sobre el software. Cómo podría volver a leer una muestra sin que se distorsione el informe al momento de imprimirlo? Y sin tener que volver a calibrar.
Your vial's cap is loose or deformed?
Many crimping and de-capping problems can be solved by two very general suggestions:
Verify the cap is properly crimped
Adjust the electronic crimper/de-capper as necessary
Verify the cap is properly…
Hi Martin,
How would I go about using conditional formatting for the Area P/F field to color the field green for Pass and Red for Fail? I'm not quite sure what type of exprssion to use.
This look like it's only bad tuning parameters. Try to have another tune file for EPA. Also, you are doing dilution with EPA. You should start with the wizard method to see if you can fix this. it has the tuning parameters. You don't need to perform resolution…
This Information Applies To: Agilent MassHunter data acquisition software and MassHunter Quantitative Data Analysis Software
Issue
This article explains the difference between quantifier and qualifier ions and provides some general guidance on quantifier...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Fragment Analyzer controller software
Issue:
Over time the dynamic coating polymer on the capillaries needs to be refreshed. Failure to do so can lead to a loss of performance. Indicators...
This Information Applies To: Agilent ICP-MS (7700/7800/7850/7900/8800/8900)
Issue
Elevated doubly charged values that are greater than 3% for x-lens configuration or greater than 7% for s-lens configuration, can be caused by contamination, or an...
This Information Applies To: Agilent 6200 Series TOF and 6500 Series Q-TOF Systems
Issue
Instructions on how to correctly dilute the tune mix for 6200 Series TOF and 6500 Series Q-TOF systems depending on source type.
Background
If the Tune...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0
Issue
The Agilent MassHunter Qualitative software can generate qualitative analysis reports, which may include sample chromatogram information and sample...
This Information Applies To: Agilent MassHunter Qualitative Analysis Software version 10.0 or higher
Issue
Identifying compounds using a spectral library is essential in qualitative analysis. This article will describe how to load a data file and...
This Information Applies To: Agilent ICP-MS Systems
Issue
The Helium cell gas line on the ICP-MS may be contaminated with air and moisture in case of a tank replacement, or if the instrument was not used for an extended amount of time. Purging the...
This Information Applies To: Agilent MassHunter Quantitative Analysis Software
Issue
The Calibration Curve window tools help with setting an appropriate calibration curve for each compound in the calibration table.
The Calibration Curve window...
This Information Applies To: Agilent MassHunter Quantitative Analysis "Quant-My-Way" user interface (version B.09.00 or later)
Issue
To process the data files in the batch and obtain quantitative results, you must either create a new quantitative...
This Information Applies To: Agilent MassHunter Qualitative and Quantitative Analysis
Issue
Calculation of signal-to-noise in MassHunter is done using the same routine for both Qualitative and Quantitative Analysis. However, there are differences...