Problem solved. The format in Region settings should be set to "English (United States)", nothing else. As a result, the Decimal symbol under Numbers in Additional settings becomes a point (Anglo-Saxon decimalization) instead of a comma (Western European…
Hi Helene,
Thank you for contacting Cell Analysis Technical Support. Sodium bicarbonate cannot be used because it is a buffering agent with a very high pKa and, unless incubated in a CO2 incubator, will raise the pH of the assay medium to well above…
Great. The 219 and 502 are quite high. Make sure that the extractor lens insulator is clean, the extractor pin isn't touching the body, the brown extractor lens connector wire is connected well and not crimped, and that the transferline tip seal ceramic…
Hi AVEdsell,
Site prep clearly states that system MUST be vented, this is a snippet from the TQ site Prep Guide about Venting. I do not know waht MS you have but this is common for all :
Bonjour Fabrice,
Juste pour ajouter a l'information que Kristen a deja donne:
Si vous selectionne la langue comme Francais, vous pouvez trouver le manuel. La description, de mon cote, a l'air d'etre en anglais, mais si vous cliquer sur le lien…
To use A-PLUS you must connect your Agilent product to a PC with a USB port using a Moxa adapter. Moxa is adapter/converter requiring specific drivers to be installed on your PC. It's important to download and install the drivers for your Moxa adapter…
Hello lperry ,
For Cool on Column inlets, the septum and the inserts types are dependent on the column size. It is quite possible that either the septum, the insert, or both, are not matching the column size.
Are the needles getting bent too? Does…
Performing a GC/MS Manual Injection Using MassHunter Software
Multiple Results
This Information Applies To: Agilent GC/MS systems using MassHunter software
Issue
Manual injections are needed for a GC/MS without an automated sampler. They are also used for troubleshooting specific issues such as no peaks, lower peaks than expected...
This Information Applies To: Agilent GC systems with Gas Clean Filter
Issue
If you have determined that the Gas Clean Filter is saturated, it mus be replaced (see the article When to Replace the Gas Clean Carrier Gas Filter ).
The steps described...
This Information Applies To: Agilent GC Systems 8890/8860/Intuvo 9000 /7890/7820/6850/6890
Issue
General guidance on GC column flow rates, outlining the correct column flow rate for optimal compound separation depends on the column internal diameter...
This Information Applies To: Agilent GC Method Translator
Issue
Detailed instructions on translating methods using the GC Method Translator.
Background
For ideal use of cases and applications with the GC Method Translator, see the article...
This Information Applies To: Agilent Solvent Vent Calculator
Issue
The Solvent Vent Calculator estimates the Agilent solvent vapor exit (SVE) accessory valve timing based on solvent type, oven temperature, vent flow, and pressure.
Background...
This Information Applies To: Use of the Agilent GC Method Translator
Issue
How to correctly use the GC Method Translator to convert a method to suit a different GC configuration or to optimize a method.
Background
The GC Method Translator...
This Information Applies To: GC Pressure Flow Calculator
Issue
Instructions for the correct use of the Pressure Flow Calculator.
Background
The Pressure Flow Calculator uses the column dimensions, oven temperature, and detector outlet pressure...
This Information Applies To: The Agilent GC Vapor Volume Calculator.
Issue
The Vapor Volume Calculator determines the vapor volume based on the injection volume, solvent type, inlet head pressure, and the inlet temperature. This calculated vapor...
This Information Applies To: Agilent GCs
Issue
Agilent provides free GC calculators and tools to assist with method development.
Background
Tips : The calculators are available for downloading at Agilent.com .
Vapor Volume...
This Information Applies To: Agilent GC Systems
Issue
The Agilent Gas Clean Carrier Gas Filter delivers faster stabilization times for increased GC productivity, removes oxygen, moisture, and hydrocarbons from the carrier gas for MS applications...