Chromatography Software Articles OpenLab ChemStation Autostart Macro Failed Error
  • Chromatography Software Articles
  • Chromatography Software Videos
  • Chromatography Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Chromatography Data Systems
    • +OpenLab CDS Administration
    • +OpenLab CDS Acquisition
    • +OpenLab CDS Data Analysis
    • -OpenLab CDS Troubleshooting
      • OpenLab ChemStation Not Starting - Basic Troubleshooting
      • OpenLab ChemStation Autostart Macro Failed Error
      • Configuration System Failed to Initialize. ChemStation Will Shutdown Now
      • License Error “The Desired Vendor Daemon is Down”
      • OpenLab CDS Online Signals Partial or Missing Display During Acquisition
      • How to Restart Agilent OpenLab Services in Windows 10
      • How to Export OpenLab System Activity Log and Diagnostics Logs for Troubleshooting
      • General Procedure and Common Issues with Licensing of Agilent OpenLab CDS

OpenLab ChemStation Autostart Macro Failed Error

Created by Linh Dinh Linh Dinh 5 months ago | Last modified by Carlos Vargas Carlos Vargas 5 months ago

This Information Applies To: Agilent OpenLab ChemStation Edition Software (C.01.xx) 


Issue

Agilent OpenLab ChemStation Edition software fails to launch with error containing the words "Autostart Macro failed”. This article will provide you some basic steps in troubleshooting the issue.


Resolution

Scenario 1. Autostart Macro failed, error #xxxx occurred. The reported error numbers are 999, 42, or 17923.

  1. Close all online and offline sessions of OpenLab ChemStation.
  2. Rename configuration files:
    • For Windows 10: C:\ProgramData\Agilent Technologies\Chemstation\# (where # is the instrument number) (Figure 1)
    • For Windows XP/Windows 7: C:\Chem32\# (where # is the instrument number)
      • CONFIG.REG to CONFIG.REG.old
      • CONF_OFF.REG to CONF_OFF.REG.old
      • ChemStationPerInstrumentSettings.config to ChemStationPerInstrumentSettings.config.old
      • instrument.config to instrument.config.old
      • RapidControl.InstrumentConfig.xml to RapidControl.InstrumentConfig.xml.old


        Figure 1: configuration files in C Drive

  3. Reconfigure the instrument by following the instruction at How to Configure an HPLC Instrument in the Agilent OpenLab Control Panel Using Auto Configuration or How to Add a GC Instrument in Agilent OpenLab CDS (Ver 2.x) via OpenLab Control Panel. Make sure that the instrument is turned on.  
  4. If Configure Instrument option in Control Panel is unavailable, ChemStation running in the background must be forced to close.
    • C:\Program Files (x86)\Agilent Technologies\ChemStation\CORE
    • Double click the CSTools.exe and click Kill ChemStation. (Figure 2)
 Tip 
Tips: File Name Extension will need to be enabled from the File Explorer View menu in order to see CSTools.exe.


Figure 2: CSTools.exe in CORE folder

Scenario 2. OpenLab ChemStation would open after acknowledging this error "Autostart Marco failed, error #65535", but all the instrument modules showed offline and, some modules were not displayed even though they were detected during the autoconfiguration.

This issue is linked to a bad global.config file that had been corrupted during the PC reboot. 

  1. Close all online and offline sessions of OpenLab ChemStation. The computer can be rebooted first to ensure that all subprocesses are shut down.
  2. Navigate to C:\ProgramData\Agilent Technologies\ChemStation\GlobalSettings (Figure 3)
  3. Rename the global.config file to global.config.old. A new global.config file will be created from the default settings.
  4. Reconfigure the instrument by following the instruction at How to Configure an HPLC Instrument in the Agilent OpenLab Control Panel Using Auto Configuration via OpenLab Control Panel. Make sure that the instrument is turned on.  


Figure 3: global.config file in GlobalSettings folder

 Tip 
Tips: Hidden items will need to be enabled from the File Explorer View menu in order to see the ProgramData folder

If the issue continues, contact your local Agilent Local Support.

 Tip 
Learn how to effectively operate your Agilent OpenLab Chromatography Data System: 
Agilent OpenLab Help & Learning Online
  • OpenLab Chromatography Data System
  • Troubleshooting
  • OpenLab ChemStation
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help