Mass Spectrometry Software Articles Calculating the Final Concentration in MassHunter Quantitative Analysis
  • Mass Spectrometry Software Articles
  • Mass Spectrometry Software Videos
  • Mass Spectrometry Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -MassHunter Software
    • +MassHunter Acquisition
    • +MassHunter Qualitative Data Analysis
    • -MassHunter Quantitative Data Analysis
      • Creating a New Batch in MassHunter Quantitative Analysis
      • Creating a Method in MassHunter Quantitative Analysis – Quant-My-Way user interface
      • Calculating the Final Concentration in MassHunter Quantitative Analysis
      • How to Average Calibration Responses in MassHunter Quantitative Analysis
      • Outlier Setup in MassHunter Quantitative Analysis Software
      • Combine multiple Methods in MassHunter Quantitative Analysis into a Single Method
      • Using the Compound Information window in MassHunter Quantitative Analysis Software
      • Modify the Integration Parameters in MassHunter Quantitative Analysis
      • Using the Batch Table in MassHunter Quantitative Analysis Software
      • Using the Calibration Curve Window in MassHunter Quantitative Analysis Software
    • +MassHunter Troubleshooting
    • +MassHunter Tips

Calculating the Final Concentration in MassHunter Quantitative Analysis

Created by Xu Li-hua Xu Li-hua 11 months ago | Last modified by Carlos Vargas Carlos Vargas 6 months ago

This Information Applies To: Agilent MassHunter Quantitative Analysis Software


Issue

The default MassHunter batch analysis reports a calculated concentration of analyte in the sample extract. The software has the capacity to apply a multiplier to the calculated concentration and report the final concentration in the original sample material.


Definitions

Calculated concentration (Calc. Conc.): The concentration after the target compound quantitated using the Calibration Table. Units are amount/mL of sample extract (e.g. μg/mL, mg/L).

Final concentration (Final Conc.): The concentration of the target compound in the sample that was extracted. It is the Calculated Concentration multiplied by a Multiplier (mg/kg or ug/L). Final Conc. = Calc. Conc.• Multiplier.

Multiplier: A calculated value that represents the sample preparation steps as per this equation:

Multiplier = Total extract volume (L) / Sample Amount (kg) • Dilution

Where:

Total sample amount (Tot.Amt) is final extract volume (L).

Sample amount (Amt.) is the mass or volume of sample portion analyzed (g or mL).

Dilution (Dil.) is the dilution factor applied to the sample extract. Expressed as 5:1 = 5.

Steps to follow

  1. See How to Create Quantitative Analysis Method in Agilent MassHunter Quant-My-Way Quantitative Analysis for calculated concentration (Calc. Conc.).
  2. Right click the top bar of the Sample section of the Batch Table to open this dialog box: Select Add/Remove Columns to open the Columns dialog box.
  3. In the Columns dialog box, under Select Columns From: drop down menu, select Sample. From the Available Columns drop down menu select Dil. > Add; Amt. > Add; Tot. Amt. > Add.
  4. In the Columns dialog box, under Select Columns From: drop down menu, select Compound Method. From the Available Columns select Multiplier > Add.
  5. In the Batch Table enter sample data for:
    1. Amt. column with the mass or volume of sample extracted.
    2. Total Amt. column with the final extract volume.
    3. Dil. column with any dilution applied to the sample extract.
  6. The Multiplier will be automatically calculated from the entered sample data.
  7. The Final Conc. will be automatically calculated from the Calc. Conc. and the Multiplier value.

Examples

Example 1: 2.567 g of a solid sample (Amt.) was extracted into 10 mL of solvent (Tot. Amt). There was no dilution made to the sample extract. Figure 1 is the sample batch report for this example.


Figure 1. Sample batch report for example 1

The instrument is calibrated in milligrams of per liter (e.g. 50 mg/L). Calc. Conc. is reported in mg/L.

  1. The Batch Table is populated with:
    1. The Total Amt. column with the final extract volume in L: 0.01 L
    2. Amt. column with the mass or volume of sample extracted in kg: 0.002567 kg
    3. The Dil. column. No dilution applied. 1:1=1
  2. The Calc. Conc. is 14.5814 mg/L of analyte.
  3. The multiplier is calculated:
    1. Multiplier = Sample extract volume (L) / Sample Amount (kg) • Dilution
    2. Multiplier = 0.01 / 0.002567 • 1 = 3.896
  4. The Final Conc. is calculated:
    1. Final Conc. = Multiplier Calc. Conc.
    2. Final Conc. = 3.896 • 14.5814 = 56.8034 mg/kg

Example 2: 2.567 g of a solid sample (Amt.) was extracted into 10 mL of solvent (Tot. Amt). The extract was diluted 5:1. Figure 2 is the sample batch report for this example.


Figure 2. Sample batch report for example 2

The instrument is calibrated in milligrams of per liter (e.g. 50 mg/L). Calc. Conc. is reported in mg/L.

  1. The Batch Table is populated with:
    1.  The Total Amt. column with the final extract volume in L: 0.01 L.
    2. Amt. column with the mass or volume of sample extracted in kg: 0.002567 kg.
    3. The Dil. column with the dilution applied to the sample extract. 5:1=5
  2. The Calc. Conc. is 14.5814 mg/L of analyte.
  3. The multiplier is calculated:
    1. Multiplier = Sample extract volume (L) / Sample Amount (kg) • Dilution.
    2. Multiplier = 0.01 / 0.002567 • 5 = 19.478
  4. The Final Conc. is calculated:
    1. Final Conc. = Multiplier Calc. Conc.
    2. Final Conc. = 19.478 • 14.5814 = 284.0168 mg/kg

 Tip 
Learn how to effectively operate your Agilent MassHunter Software:
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 6400 Series Triple Quadrupole LC/MS with MassHunter Workstation
Agilent 7000 or 7010 Triple Quadrupole GC/MS with MassHunter Workstation e-learning paths available from Agilent education 
  • Operation
  • gcms data analysis
  • Quantitative Data Analysis Software
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help