Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Mass Spectrometry Software Portal
  • Agilent Knowledge Portal
Mass Spectrometry Software Portal
Articles Calculating the Final Concentration in MassHunter Quantitative Analysis
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Mass Spectrometry Software Portal Home
  • Mass Spectrometry Software Portal Articles
  • Mass Spectrometry Software Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  7 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Mass Spectrometry Software
  • Component :  GC/MS Data Analysis Software LC/MS Data Analysis Software
  • Product Name :  MassHunter Quantitative Analysis Software
  • Task :  Data Analysis Operation
Related Articles
  • Creating a Method in MassHunter Quantitative Analysis "Quant-My-Way" User Interface

    This Information Applies To: Agilent MassHunter Quantitative Analysis "Quant-My-Way" user interface (version B.09.00 or later) Issue To process the data files in the batch and obtain quantitative results, you must either create a new quantitative method or use a previous quantitative method and then apply it to the batch. Steps to follow 1. Create a new batch and add the applicable data files (Add Samples) to begin the quantitative analysis then initially save the batch. Refer to How...
  • How to Match the Signal-To-Noise Calculations between MassHunter Qualitative and Quantitative Analysis software for Extracted Ion Chromatograms

    This Information Applies To: Agilent MassHunter Qualitative and Quantitative Analysis Issue Calculation of signal-to-noise in MassHunter is done using the same routine for both Qualitative and Quantitative Analysis. However, there are differences in the way that the parameters are input. To get the same value between the programs, the user must make sure that they are inputting the same parameters. The following procedure is to calculate the signal-to-noise and get the same value in the Qualitative...
  • Outlier Setup in MassHunter Quantitative Analysis Software

    This Information Applies To: Agilent MassHunter Quantitative Analysis Software Issue MassHunter Quantitative Analysis allows limits to be set on many result values, and for results outside these limits (outliers) to be flagged in the batch report. The flagged results are easily identified, enabling fast visual identification of outliers, which improves productivity and, reduces the risk of reporting invalid data. The following procedure introduces the use of outliers within the MassHunter...
  • How to Match the Signal-To-Noise Calculations Between MassHunter Qualitative and Quantitative Analysis Software for Total Ion Chromatograms

    This Information Applies To: All versions of Agilent MassHunter Qualitative Analysis Software , and Agilent MassHunter Quantitative Analysis version B08.00 and up Issue Calculation of signal-to-noise in MassHunter is performed using the same parameters for both Qualitative and Quantitative Analysis, however the parameters are entered with different steps. To match the results for both programs, the user must make sure to input the same parameters. The following procedure shows how to calculate...
  • Creating a New Batch in MassHunter Quantitative Analysis

    This Information Applies To: Agilent MassHunter Quantitative Analysis Software Classic mode, and Agilent MassHunter Quantitative Analysis Software Quant-My-Way mode (Rev. B.09.00 and above) Issue In the MassHunter quantitative analysis, you should begin with creating a new batch file. The following is the procedure to create a new batch. Steps to follow For the MassHunter Quant-My-Way Quantitative software 1. To create a new batch, click File or New Batch (see Figure 1 ). Figure 1...

Calculating the Final Concentration in MassHunter Quantitative Analysis

Answer

This Information Applies To: Agilent MassHunter Quantitative Analysis Software


Issue

The default MassHunter batch analysis reports a calculated concentration of analyte in the sample extract. The software has the capacity to apply a multiplier to the calculated concentration and report the final concentration in the original sample material.


Definitions

Calculated concentration (Calc. Conc.): The concentration after the target compound quantitated using the Calibration Table. Units are amount/mL of sample extract (e.g. μg/mL, mg/L).

Final concentration (Final Conc.): The concentration of the target compound in the sample that was extracted. It is the Calculated Concentration multiplied by a Multiplier (mg/kg or ug/L). Final Conc. = Calc. Conc.• Multiplier.

Multiplier: A calculated value that represents the sample preparation steps as per this equation:

Multiplier = Total extract volume (L) / Sample Amount (kg) • Dilution

Where:

Total sample amount (Tot.Amt) is final extract volume (L).

Sample amount (Amt.) is the mass or volume of sample portion analyzed (g or mL).

Dilution (Dil.) is the dilution factor applied to the sample extract. Expressed as 5:1 = 5.

Steps to follow

  1. See How to Create Quantitative Analysis Method in Agilent MassHunter Quant-My-Way Quantitative Analysis for calculated concentration (Calc. Conc.).
  2. Right click the top bar of the Sample section of the Batch Table to open this dialog box: Select Add/Remove Columns to open the Columns dialog box.
  3. In the Columns dialog box, under Select Columns From: drop down menu, select Sample. From the Available Columns drop down menu select Dil. > Add; Amt. > Add; Tot. Amt. > Add.
  4. In the Columns dialog box, under Select Columns From: drop down menu, select Compound Method. From the Available Columns select Multiplier > Add.
  5. In the Batch Table enter sample data for:
    1. Amt. column with the mass or volume of sample extracted.
    2. Total Amt. column with the final extract volume.
    3. Dil. column with any dilution applied to the sample extract.
  6. The Multiplier will be automatically calculated from the entered sample data.
  7. The Final Conc. will be automatically calculated from the Calc. Conc. and the Multiplier value.

Examples

Example 1: 2.567 g of a solid sample (Amt.) was extracted into 10 mL of solvent (Tot. Amt). There was no dilution made to the sample extract. Figure 1 is the sample batch report for this example.


Figure 1. Sample batch report for example 1

The instrument is calibrated in milligrams of per liter (e.g. 50 mg/L). Calc. Conc. is reported in mg/L.

  1. The Batch Table is populated with:
    1. The Total Amt. column with the final extract volume in L: 0.01 L
    2. Amt. column with the mass or volume of sample extracted in kg: 0.002567 kg
    3. The Dil. column. No dilution applied. 1:1=1
  2. The Calc. Conc. is 14.5814 mg/L of analyte.
  3. The multiplier is calculated:
    1. Multiplier = Sample extract volume (L) / Sample Amount (kg) • Dilution
    2. Multiplier = 0.01 / 0.002567 • 1 = 3.896
  4. The Final Conc. is calculated:
    1. Final Conc. = Multiplier Calc. Conc.
    2. Final Conc. = 3.896 • 14.5814 = 56.8034 mg/kg

Example 2: 2.567 g of a solid sample (Amt.) was extracted into 10 mL of solvent (Tot. Amt). The extract was diluted 5:1. Figure 2 is the sample batch report for this example.


Figure 2. Sample batch report for example 2

The instrument is calibrated in milligrams of per liter (e.g. 50 mg/L). Calc. Conc. is reported in mg/L.

  1. The Batch Table is populated with:
    1.  The Total Amt. column with the final extract volume in L: 0.01 L.
    2. Amt. column with the mass or volume of sample extracted in kg: 0.002567 kg.
    3. The Dil. column with the dilution applied to the sample extract. 5:1=5
  2. The Calc. Conc. is 14.5814 mg/L of analyte.
  3. The multiplier is calculated:
    1. Multiplier = Sample extract volume (L) / Sample Amount (kg) • Dilution.
    2. Multiplier = 0.01 / 0.002567 • 5 = 19.478
  4. The Final Conc. is calculated:
    1. Final Conc. = Multiplier Calc. Conc.
    2. Final Conc. = 19.478 • 14.5814 = 284.0168 mg/kg

 

 Tip 
Learn how to effectively operate your Agilent MassHunter Software:
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 6400 Series Triple Quadrupole LC/MS with MassHunter Workstation
Agilent 7000 or 7010 Triple Quadrupole GC/MS with MassHunter Workstation e-learning paths available from Agilent education 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!