This Information Applies To: Agilent LC Pumps with built-in Degasser
Issue
If the built-in degasser is not working and gives an error on power-on, the pump will be in error mode, and therefore no injections can be made. The user may wish to disable [or enable] the degasser, especially in an isocratic run, to be able to still use the pump while the degasser is repaired.
This article details how to disable the degasser.
Steps to follow
- Launch Lab Advisor and connect to the Instrument.
- Go to Instrument Control > [Pump model] > Controls > Control > Degasser Controls and select Degasser > Off. (See Figure 1)
Note: If you turn off the degasser, and cycle power the pump, it will revert to being back on so will need to repeat the preceding steps each time.
Figure 1. Built-in degasser Lab Advisor control
|
Tips: For additional information on degasser troubleshooting, see: Troubleshooting "Vacuum Limit Not Reached" Error in Agilent LC Pumps
|