ICP-OES Articles Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • +ICP-OES Maintenance
    • -ICP-OES Troubleshooting
      • ICP-OES Torch Overheating
      • Software Status Indicators 5000 Series ICP-OES
      • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections
      • Checking for a Blocked ICP-OES Nebulizer
      • Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
      • How to Power Cycle Agilent 5000 ICP-OES Systems
      • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
      • Agilent ICP-OES Performance Test
      • Bypassing the AVS for the 5000 Series ICP-OES
      • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
      • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Created by Tina Harville Tina Harville over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 17 days ago

This Information Applies To: Agilent G3292A and G8481A Chillers used on ICP-MS and ICP-OES instruments


Issue 

When experiencing issues with your Agilent chiller, please try a factory reset first, as shown in the following steps and in Figure 1.


Resolution 

1. Place the main power switch on the rear of the instrument in the off position.

2. Press and hold the soft power button on the front panel while returning the main power switch to the on position (O).


Figure 1. Performing a Factory Reset of the Chiller

After performing a factory reset, you will need to set the chiller temperature to 20 °C:

1. Press the Select/Set knob on the front panel (right above the soft power button).

2. The decimal point at the bottom right of the display will flash, indicating that the temperature can be changed.

3. Rotate the Select/Set knob until the desired setpoint temperature (20 °C) is displayed.

4. The setting is accepted after the Select/Set knob is pressed a second time or automatically after a few seconds of inactivity.

If the factory reset doesn't help, and you are getting a "Lo H2O" error on the front display of the chiller, please check the fuses after ensuring that the chiller is full of coolant:

 Warning 
Warning: Always unplug the chiller before removing any covers.

1. Unplug the chiller.

2. From the back of the chiller, slide off the top cover after removing the two screws on the left and right side, as shown in Figure 2 and Figure 3:


Figure 2. Remove the screws at the top on the back of the chiller.


Figure 3. Remove the top cover of the chiller.

3. Then remove the two screws at the top of the front of the chiller, as shown in Figure 4.


Figure 4. Remove the screws at the top of the front of the chiller.

4. Once you have access to the main board, identify the two fuses, as shown in Figure 5.

 
Figure 5. The two fuses on the main board of the chiller.

5. Remove and check the continuity of the two fuses with a multi-meter.

6. If new fuses are needed, they can be obtained at any home improvement store.

7. Replace the bad fuses, and reassemble the screws and the top cover.

8. Turn on the main power to the chiller.

 Tip 
Learn how to effectively troubleshoot your ICP-OES instrument:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education

 

  • ICPOES
  • G8481A
  • G3292A
  • ICPMS
  • chiller fuses
  • Agilent Support Insights
  • ICPOES chiller
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help