This Information Applies To: Agilent demountable torches
Issue
The Agilent 5000 series ICP-OES instruments can be fitted with a demountable, semi-demountable, or bonded torch. The demountable and semi-demountable torches can be disassembled for easy cleaning. The demountable torch allows quick...
This Information Applies To: 5100, 5110, 5800, and 5900 ICP-OES instruments
Issue
The ICP-OES operator can identify the source of plasma ignition failure. The main cause is air entering the plasma.
Steps to follow
Excessive air entering the plasma. Check the following:
The peristaltic...
This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900)
Issue
How to run the ICP-OES Performance Test after installation, preventative maintenance, repairs, and when troubleshooting.
Steps to Follow
The ICP-OES Performance Test runs a suite of tests including Resolution...
This Information Applies To: Agilent 5000 Series ICP-OES
Issue
Blocked nebulizers are the source of many instrument performance issues including loss of sensitivity and poor precision, especially when running organics, high solids, or high salt concentration solutions. The ICP-OES backpressure...
This Information Applies To: Agilent VGA 77 (Vapour Generation Accessory)
Issue: This article discusses the setup steps for the VGA 77.
VGA 77
Warning : Noxious Fumes and Corrosive Liquids. Vapor generation methods of analysis involve the generation of toxic hydrides and require the...