Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  1 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Systems
  • Component :  GC Inlets
  • Product Name :  6850 GC System 6890 GC System 7820 GC System 7890 GC System 8860 GC System 8890 GC System Intuvo 9000 GC System
  • Task :  Troubleshooting
Related Articles
  • Replacing the Inlet Septum for Split/Spitless and Multimode Inlets

    This Information Applies To: Agilent Intuvo 9000, 8860, 8890, 7890A/B, 7820A, 6890, 6850 GC Systems with Split/Splitless, or Multimode Inlets. Issue: This article describes how to replace the inlet septum for Split/Splitless and multimode inlets for all Agilent GC Systems. Regular replacement of the septum can reduce small leaks in the inlet, allow for more reproducible chromatography, and protect the column from degradation. Replacement intervals should be every 100 to 400 injections depending...
  • How to Calibrate the Agilent 7693A Sample Tray

    This Information Applies To: Agilent 7693A Automatic Liquid Sampler Issue: The Automatic Liquid Sampler (ALS) system calibration process aligns the sample tray with the injector turret position to enable vial transfers without incident. Calibration should be performed if a calibration does not exist, as a routine maintenance procedure and is recommended if any ALS components are moved. Do not confuse ALS alignment, with ALS calibration. During alignment, the ALS tray gantry will basically...
  • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC

    This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet Issue Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet degrades over time due to accumulated contamination (e.g., baseline noise, ghost / artifact peaks, loss of reproducibility, carry over, loss of compound response or no compound response). Items Required The following Gold seals are available...
  • Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System

    This Information Applies To: Agilent 8890/8860 GC Systems with SSL, MMI, PTV, or VI inlets. Issue Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure is too high, or the split vent flow cannot be reached. The replacement frequency will vary with the GC application. For other GC models, see: Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System Items Required QuickPick...
  • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems

    This Information Applies To: Agilent 7890 A/B, 8860 and 8890 GC Systems with Split/Splitless (SSL), Multimode (MMI) or Programmable Temperature Vaporization (PTV) Inlets Issue: The Inlet Leak & Restriction Test is executed after every inlet maintenance procedure. It can also be run for troubleshooting purposes or as a frequent confidence check for GC performance. Useful to Find and repair medium to large leaks. Useful when the inlet will not pressurize to the setpoint due to a large leak...

Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint

Answer

This Information Applies To: Agilent GC systems, focusing on 8890 and 7890A/B, but also applies to 6850A/B, 6890A/N, 7820A, and 8860.


Issue:

If the inlet pressure / carrier gas flow fail to achieve the setpoint after a few minutes, the inlet pressure will shut down automatically. This article will help you to identify the possible causes of this issue.


Steps to follow: 

  1. Leak at the inlet
    To detect big leaks in the inlet, perform one of the diagnostic tests for the inlet: Inlet Leak Check, or Leak and Restriction Test depending on the instrument model. (How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems) If the results of the test show an important leak, check the following causes
    1. Septum hole is deformed or widened. Installing a new septum will resolve the issue.

    2. Inlet to column connection is not fully tight, wrong positioned or column is broken near the inlet. Recommended to check for breaks, trim the column if necessary, use new ferrules, and reinstall the column. See capillary column installation guide.

    3. Liner, O-ring, split filter, gold seal, or washer under gold seal are not positioned correctly or damaged. Exchange them if you find them fractured or broken.

    4. Make sure the fittings around the inlet and the weldment nut had been tightened properly. Check if tubes at weldment are intact.

       Tip  Tips: For further instructions see Troubleshooting a Leak in the Split/Splitless Inlet , Inlet Diagnostics 6890/7890, 8890 Troubleshooting Manual, 8860 Troubleshooting Manual.
  2. Source pressure not high enough

    1. The carrier gas regulator must be set at least 20 psi greater than the maximum inlet pressure required in the GC method. Check the Site Preparation Guide of your respective GC model to get more information on gas supply pressure requirements.

    2. Make sure the gas regulator on/off valve is fully opened.

    3. Is the gas cylinder full or going to finish? Agilent recommends exchanging the gas cylinder when its pressure is about 150 psi. For more information, watch the video about Gas Management Best Practice.

       Tip 
      Tips: 8890 and 8860 GCs have a Gas Supply Pressure Check diagnostic tool. This tool is not available if a mass selective detector is configured with the system.
    4. Run the Gas Supply Pressure Check in the 8890 or 8860 GC. On the touch display, press Diagnostics > Diagnostic Tests > Front (or Back) Inlet > Gas Supply Pressure Check > Start Test (See Figure 1).
      The test checks for the maximum pressure the gas supply can deliver. It will pass if the gas supply can deliver at least 276 kPa (40 psi) to the instrument.


       
       
      Figure 1. Gas Supply Pressure Check for the Agilent 8890 GC

  3. Configuration

    1. Check if a column is installed in the inlet.

    2. Compare installed and configured column type. Configured column length and diameter will affect the calculation of the inlet gas pressure to set the specified column flow.

    3. Check if the correct gas type is connected and configured.

    4. Make sure that the split flow or gas saver flow is at least 15 ml/min greater than the column flow.

If you still cannot determine where the leak is located after have been following the preceding steps, contact your local Agilent support representative.

 Tip 
Learn how to effectively troubleshoot your split/splitless inlet:
Agilent 7890 Series GC: Split/Splitless Inlet Troubleshooting e-learning course available from Agilent education.



 