Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Cleaning the 5000 Series ICP-OES Torch, Organic Matrices
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES
  • Task :  Maintenance
Related Articles
  • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue The ICP-OES torch plasma ionizes the introduced gas or aerosol stream. In normal operating conditions, incomplete ionization occurs and material is deposited on the torch surface. Build up of deposits can create blockages, damage the torch or interact with the sample stream causing poor precision, accuracy, and memory effects. ICP-OES overview The ICP-OES sample introduction components are detailed in...
  • Can the Agilent 2100 Bioanalyzer be Used to Determine Size of RNA?

    This Information Applies To: 2100 Bioanalyzer, RNA 6000 Nano kit, RNA 6000 Pico kit, Small RNA kit Issue: Is it possible to size RNA on the Agilent 2100 Bioanalyzer system? Resolution: Agilent offers total and small RNA assays on the Bioanalyzer. There are no sizing specifications available for total RNA assays as the assays are not developed or validated for sizing accuracy and precision. However, given that ladder consists of RNA fragments from 200 nt to 6000 nt, it is possible to...
  • SIPS 10/20 Maintenance

    This Information Applies To: Agilent AA240, AA280 Atomic Absorption Systems with Sample Introduction Pump System (SIPS) Accessory Issue The following maintenance is recommended for the SIPS Accessory, model 10 (single pump), or model 20 (dual pump). SIPS 10/20 Accessory SIPS 20 Connections 1 Sample, 2 Standard or modifier (SIPS 20 only), 3 Diluent, 4 To nebulizer, 5 Continuous flow See SIPS 20 Video to watch the SIPS 20 operating with SpectrAA Software Steps to follow ...
  • Peak Fronting in NGS Library Smears on Agilent TapeStation

    This Information Applies To: 2200/4200/4250 TapeStation systems, TapeStation D1000 ScreenTape assay, TapeStation High Sensitivity D1000 ScreenTape assay, TapeStation analysis software Issue: A sharp peak appears at the lower end of the NGS library smear in Agilent TapeStation D1000 ScreenTape and High Sensitivity D1000 ScreenTape assays (see Figure 1 ). Figure 1 . Region analysis (blue borders) of an NGS library smear. The red arrow indicates the peak fronting. Resolution: Peak...
  • Flame AA Burner Maintenance

    This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280) Issue: This article describes when and how to clean an AA burner. Background For optimum and safe performance, burners (see Figure 1 ) should be cleaned after each day of use. Particularly if the solutions being analyzed have a high solids content or a rich nitrous oxide-acetylene flame has been used. If not cleaned regularly, hard deposits that are difficult to remove may build up...

Cleaning the 5000 Series ICP-OES Torch, Organic Matrices

Answer

This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900)


Issue

The ICP-OES torch plasma ionizes the introduced gas or aerosol stream. In normal operating conditions, incomplete ionization occurs and material is deposited on the torch surface. Build up of deposits can create blockages, damage the torch or interact with the sample stream creating poor precision, accuracy, and memory effects.


ICP-OES overview

The ICP-OES sample introduction components are detailed in Figure 1.

 

Figure 1: ICP-OES sample introduction components

1. Torch loader handle 2. Torch 3.Torch clamp 4. Nebulizer gas connections 5. Spray chamber drain tubing 6. Spray chamber 7. Nebulizer 8. Nebulizer inlet tubing

 Caution 
Caution: The torch is fragile. Do not use cleaning wire or abrasives like brushes or a scourer to clean the torch. Avoid acid contact with the seal where the quartz meets the plastic base. Damage to the seals and torch body may occur.

Steps to follow

Removing the torch:

  1. Turn off the plasma and allow instrument to cool.

     Caution 
    Caution: The torch and torch compartment become hot during instrument operation and remain so for some time after the plasma has been switched off. Allow the plasma compartment to cool before accessing the plasma compartment. Allow the torch to cool before touching the outer tube or attempting to dismantle the torch.
  2. Disconnect the spray chamber (Figure 1).

  3. Remove the torch from the instrument.

  4. Disassemble fully demountable torches (Figure 2) and semi-demountable torches (Figure 3).

    Figure 2: Fully demountable torch

    Figure 3: Semi-demountable torch

  5. Inspect the torch components for damage including holes or cracks in the plastic housing, seal or glass tubing, and the outer quartz tube becoming rough to touch. Discoloration of the plastic housing occurs with long-term exposure to acid and does not impact on performance. Replace as necessary.

Cleaning the torch:

 Caution 
Caution: Using the muffle furnace or portable handheld propane torch will heat the torch to high temperatures. Always wear appropriate Personal Protective Equipment.
  1. Fully and semi-demountable torches: Place only the quartz ICP torch tubes in a muffle furnace at approximately 500oC for one hour. Do not put plastic components in the muffle furnace. 
  2. One piece torch and quartz components attached to plastic torch body
    1. Do not remove the torch adhesive cap on the one-piece torch that covers the seal between the quartz and the plastic.
    2. Wrap a wet cloth around the plastic components. 
    3. Use a hand-held propane torch to heat the quartz and remove deposits.
  3. The outer tube and injector should be stood in a vertical position until they have cooled.

Rinsing the torch:

  1. Hold the torch with the ball joint connector at the top and thoroughly rinse the outside, inside through ball joint connector, upper and lower gas ports.

  2. Rinse the removable top seal on semi-demountable torches.

Drying the torch

  1. Flush all torch components with laboratory grade compressed air or nitrogen. Flush the torch through the gas supply ports and the ball joint connector, then invert and flush through the open end of the quartz outer tube set. Do not insert the air/nitrogen nozzle inside the tube set. When flushing is not available, dry glass components in a drying oven and plastic components with a lint free cloth.

  2. Ensure that all moisture is removed then re-assemble and refit the torch.

  3. When not is use, store the torch in the original packaging or the ICP-OES Easy-fit torch storage rack (part number G8010-67000).

Additional checks

  1. When refitting the torch, inspect for damage such as loose fitting of the quartz tubes in the plastic base, holes, or significant cracks. If any damage is found, replace the torch immediately.

  2. Check for carry-over after refitting the torch to determine if the cleaning procedure has been sufficient. If carry-over is found, repeat the cleaning process.

  3. Replace the torch when the outer surface of the quartz outer tube is rough to the touch, which indicates signs of wear, devitrification, or if there are any cracks visible.

 Tip  Recommended maintenance: Agilent recommends daily inspection and weekly cleaning of the torch. Analysis of samples with complex matrices may require more frequent cleaning.

 Tip 
Learn how to effectively maintain you ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help