This Information Applies To: ICP-MS MassHunter 4.x/5.x
Issue
Reviewing the latest Performance Report, and Performance Report History is a useful tool for troubleshooting ICP-MS instrument performance issues. Historical sensitivity, background, precision, hardware settings, tune parameters, and...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to perform routine maintenance on the Agilent ICP-MS Torch.
Background
Deposits on the torch can be removed by routine maintenance. Maintenance frequency depends on sample throughput and sample type...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
The torch, bonnet, and platinum shield plate must be installed after performing maintenance, or replacing with new components.
Items Required
Standard Torch, quartz, 2.5 mm id - not suitable for samples...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
The torch, bonnet, and shield plate must be removed and inspected during routine maintenance, and when troubleshooting plasma ignition problems. Faults with the torch, bonnet, or shield plate can cause plasma...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
This procedure tests for an air leak in the sample introduction system, or fault with spray chamber draining, by sealing the end of the torch with a piece of parafilm and testing plasma ignition. Sealing with...