This Information Applies To: Agilent ICP-MS MassHunter Software version 5.1 and above
Issue
The user manuals for Agilent ICP-MS Systems for MassHunter 5.1 and above are included in the ICP-MS MassHunter Help and Learning Center. There are manuals for the software, hardware, optional software, and...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination, or a fault with the sample introduction...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
What information does the ICP-MS instrument status indicator provide?
Background
The ICP-MS Instrument Status Indicator on the instrument, and in the ICP-MS MassHunter Software is useful to assess the...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power off Agilent ICP-MS systems before maintenance, for long-term shutdown, or for power cycling the instrument. Power cycling may be required when troubleshooting communication, lighting the plasma...
This Information Applies To: Agilent 7700, 7800, 7850 and 7900 series ICP-MS systems
Issue
Over time the water inlet strainer can become clogged and reduce water flow. Reduced water flow limits the system cooling and can result in the instrument shutting down during analysis and triggering error...