Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
GC/MS Articles How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • +GC/MS Operation
    • -GC/MS Maintenance
      • How to Open the Side Plate of the GC/MSD
      • How to Clean the EI/CI Source in the Agilent 5977A/B
      • How to Clean the HES (High Efficiency Source)
      • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)
      • How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
      • Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
      • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the CI Source
      • How to Disassemble and Assemble the HES (High Efficiency Source)
      • How to Bake Out an Agilent GC/MSD
    • +GC/MS Troubleshooting

How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources

Created by Agilent Knowledge Portal Agilent Knowledge Portal 6 months ago | Last modified by Agilent Knowledge Portal Agilent Knowledge Portal 3 months ago

This Information Applies To: Agilent EI Source


Issue:

During normal operation of the GC/MS, the EI SST/Inert ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filaments and ceramic insulators). The ion source will then need to be assembled prior to re-installing. Correct abrasive cleaning, assembly, and installation are essential to robust and reliable operation.


Tools Needed:

• 597X Tool Kit and Cleaning Supplies p/n G7077-60566 (contains)

  • Gloves, clean, lint-free (8650-0030)
  • Cloths, Lint Free (05980-60051)
  • Nut Driver 5.5mm (8710-1220)
  • Tweezers (8710-0007) 
  • Hex Driver 1.5mm (8710-1570)
  • Wrench 10mm Open End 

Figure 1: Parts Breakdown of the EI Source

Item Number Item Description Part Number
1 Gold plated set screw G1999-20022
2 Gold plated screw G3870-20021
3 Interface socket G1099-20136
4 Source body G1099-20130
5 Drawout cylinder G1072-20008
6 Drawout plate, 3 mm 05971-20134
7 4-turn filament G7005-60061
8 Spring washer 3050-1374
9 Flat washer 3050-0982
10 Lens insulator G3170-20530
11 Entrance lens G3170-20126
12 Ion focus lens 05971-20143
13 Repeller insulator G1099-20133
14 Repeller G3870-60172
15 Flat washer 3050-0982
16 Belleville spring washer 3050-1301
17 Repeller nut 0535-0071
18 Source heater block assembly* G3870-60180
19 Repeller block insert G3870-20135

* If replacement is required then contact Agilent for repair options

To Disassemble the EI source:

  1. Remove the ion source and place it on a clean cloth
  2. Use the 1.5mm Hex Driver to remove the gold plated screws and washers securing the filaments to the ion source. ( See Figure 2 for steps 2-5)
  3. Remove both filaments from the ion source.
  4. Use a 1.5mm Hex Driver to remove the two screws securing the repeller / heater-sensor assembly and separate the repeller / heater-sensor assembly from the ion source body.
  5. Use the 5.5 mm Nut Driver to remove the repeller nut, then remove the washers and insulators from the source heater-sensor assembly, then remove the repeller.


    Figure 2
    . Removing the filaments, repeller and insulators from the source heater block assembly.

  6. Use the 1.5mm Hex Driver to remove the gold plated set screw from the side of the ion source body. (See Figure 3 for steps 6-10)
  7. Pull the entrance lens and ion focus lens to remove them from the ion source body.
  8. Separate the entrance lens and ion focus lens from the lens insulator.
  9. Remove the drawout plate and the drawout cylinder from the source body.
  10. Use the 10mm Wrench to remove the interface socket.


    Figure 3. Removing the lens, drawout plate, drawout cylinder and insulators from the source body. 

  11. See How to Clean the EI/CI Source in the Agilent 5977A/B or assembly if broken insulators or filaments were being replaced.

To Assemble the EI Source:

  1. Assemble the ion focus lens, entrance lens, and lens insulators. (See Figure 4 for steps 1 - 5)
  2. Install the interface socket into the source body.
  3. Insert the drawout plate and the drawout cylinder into the source body.
     Caution 
    Caution: Do not overtighten the repeller nut or the ceramic repeller insulators will break when the source heats up. The nut should only be finger-tight
  4. Slide these assembled part into the source body.
  5. Install the setscrew that holds the lenses in place.


    Figure 4. Assembling the lens, insulators, drawout plate and drawout cylinder into the source body. 

  6. Assemble the repeller assembly. ( See Figure 5 for steps 6 - 11)
  7. Install the repeller block insert into the source heater block assembly
  8.  Install the repeller insulators into the source heater block assembly and repeller block insert. 
  9. Install the repeller through the repeller insulators, then put the flat washer followed by the Belleville spring washer onto the end of the repeller shaft and secure finger tight with the repeller nut.
  10. Attach the repeller assembly to the source body using the two gold plated screws and spring washers.
  11. Install the filaments using the two gold plated screws and spring washers.
     Caution 
    Caution: Do not overtighten the interface socket. Overtightening could strip the threads.


Figure 5. Assembling the repeller, insulators, and filaments into the source heater block assembly. 

 Tip 
Learn how to effectively operate your 5977:
Agilent 5977 GC/MSD with MassHunter Workstation or Agilent 5977 GC/MSD with OpenLab CDS e-learning course on Agilent University
  • GC/MS
  • single quadrupole
  • 5977B
  • 5977A
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help