GC Articles Using the Condensation Test to Troubleshoot the GC Inlet
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • +Basic Troubleshooting for GC Systems
      • Using the Condensation Test to Troubleshoot the GC Inlet
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
      • Gathering GC Diagnostics Log Files Using the Agilent Acquisition Software
      • Gathering Diagnostics Log Files from the 8890/8860/Intuvo 9000 GC Systems Using the Browser Interface
    • +GC Best Practices

Using the Condensation Test to Troubleshoot the GC Inlet

Created by Agilent Knowledge Portal Agilent Knowledge Portal 8 months ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: Agilent GC Systems 


Issue

This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler. 


Background

Typical chromatographic features that indicate possible contamination problems include ghost peaks, carryover, excessive baseline noise or baseline drift, and instability. See Basic Troubleshooting for GC Systems .

Steps to Follow

To identify if the inlet is the source of contamination: 

  1. Leave the GC between 40 to 50 °C for eight or more hours

  2. Run a blank analysis (i.e., start the GC, but with no injection) using the normal temperature conditions and instrument settings

  3. Collect the chromatogram for this blank run

  4. Immediately repeat the blank run when the first run one is completed. Do not allow more than 5 minutes to elapse before starting the second blank run

  5. Collect the chromatogram for the second blank run and compare it to the first chromatogram. If:

    1. Both chromatograms contain few peaks or little baseline drift, the carrier gas, carrier gas lines, gas filters and inlet are relatively clean

    2. The first chromatogram contains a larger number of peaks and baseline instability, any combination of the incoming carrier gas, carrier gas line, gas filters or the inlet is contaminated. See Basic Troubleshooting for GC Systems .

       Tip 
      Tip: Gas supply problems are most likely to occur immediately after changing gas cylinders. 

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1001e - Gas Management Best Practices
GC-0GEN-1012s - GC Inlets Theory and Operation e-learning courses available from Agilent education

  • Troubleshooting
  • 7820
  • Intuvo 9000
  • GC Systems
  • 7890
  • 8890
  • 8860
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help