Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles How To Replace the Nitrogen Phosphorus Detector Bead in 7890/6890 GC Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  5 May 2023
  • Modified :  28 Jul 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Systems
  • Component :  GC Detectors
  • Product Name :  6890 GC System 7890 GC System
  • Task :  Maintenance
Related Articles
  • Capillary Column Installation Quick Reference Guide

    This Information Applies To: All Agilent Gas Chromatographs, 6890, 6850, 7890, 7820, 8890, and 8860. Issue: A quick reference guide on the capillary column installation for the GC. Inlets Detectors Split/splitless inlet Purged packed inlet Multimode Inlet Cool on Column (COC) Inlet PTV Inlet Volatiles Interface (VI) Inlet FID / NPD TCD - Standard Low Leakage Ferrules TCD - Original Design Column Adapter µECD FPD Steps to follow Warning : The...
  • Recommended Parameters for GC Maintenance Methods

    This Information Applies To: Agilent GC systems Issue Before most maintenance procedures, the GC must be set to safe temperatures, pressures, or flows. You can create and save the maintenance method in the controlling software or the GC itself. A proper maintenance method will protect the user from injures, and the instrument from damage while performing maintenance. Tips : The Intuvo 9000, 8890 and 8860 GC systems will load the right instrument setpoints while performing guided...
  • Agilent GC Detector Sampling (Data) Rate

    This Information Applies To: Agilent GC Systems Issue The data rate determines the number of sampling points across the chromatographic peak and the sample background. The data rate is based on the width of the analyte speak (in minutes) and the ratio of the analyte signal to the background noise. Background The data rate unit is hertz (Hz), where 1 Hz is one cycle per second. The data rate determines the number of data points taken across the width of the peak, and in the background...
  • Bake Out the Agilent Flame Ionization Detector

    This Information Applies To: Agilent GC Systems with flame ionization detector (FID) Issue Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces. Items required Ferrule, no hole 10/pk (p/n 5190-4054 ) Column nut, any of the following: Column nut, universal (p/n 5181-8830 ) Blanking plug, finger-tight style (p/n 5020-8294 ) Finger-tight column nut (p/n 5020-8292 ) Column nut, collared, self...
  • When to Replace the Gas Clean Carrier Gas Filter

    This Information Applies To: Agilent GC Systems Issue The Agilent Gas Clean Carrier Gas Filter delivers faster stabilization times for increased GC productivity, removes oxygen, moisture, and hydrocarbons from the carrier gas for MS applications, and provides ultimate GC column protection. The filter must be replaced before it is consumed. Background As gas passes through the filter material, the filter picks up the impurities and eventually becomes saturated over time. When saturation...

How To Replace the Nitrogen Phosphorus Detector Bead in 7890/6890 GC Systems

Answer

This Information Applies To: Agilent 7890A and 6890 Gas Chromatography Systems


Issue

Agilent Nitrogen Phosphorus Detectors (NPDs) use three different types of beads with different characteristics. The NPD method settings within the GC method are specific to the type of bead used in the NPD. Damage to the newly replaced bead can occur if incorrect method settings are used for the type of installed bead. 

This article details the process to change the bead in Agilent NPD detectors. 


Steps to follow:

Exploded view of the NPD.
Figure 1. Exploded view of the NPD
1. Bead Assembly, 2. Lid 3. Metal C-ring, top, 4. Ceramic Insulator, upper, 5. Collector, 6. Metal C-ring, bottom, 7. Ceramic Insulator, bottom, 8. Metal C-ring, bottom, 9. Jet, 10. NPD cover, 11.Column Adapter, 12. Insulation, 13. Insulation, 14. Insulation cup, 15. Ferrule, 16. Column nut, 17. Bead assembly cable, 18. NPD lid standoff, 19. J-clamp, 20. Electrometer

  1. Identify and order the correct bead for the NPD. The three types of beads are:

    • White Ceramic Bead (p/n G1534-60570)

    • Black Ceramic Bead (p/n 5183-2007)

    • Blos Bead (p/n G3434-60806)

      NPD Blos bead
      Figure 2. NPD Blos bead 

  2. Set the NPD bead voltage to 0.0. (Do not set it to Off.)

  3. Set Adjust Offset to Off. 

  4. Cool the detector to 60 °C or lower. Leave all gas flows on. To cool the detector faster, raise the GC detector cover and open the hinged NPD cover.
    Put on lint-free gloves before touching any of the detector parts.

  5.  Twist the ring to disconnect the bead assembly cable. Push and twist the lock so that the button slides up in the groove, then pull the cable ends apart.

    Disconnecting the bead assembly.
    Figure 3. Disconnecting the bead assembly 

  6. Remove the 3 T- 10 Torx screws from the bead assembly.

    Removing of the screws from the bead assembly.
    Figure 4. Removing the screws from the bead assembly

  7. Gently lift and remove the old bead assembly. Avoid bumping the bead on the sides of the collector.

    Removing of the bead assembly.
    Figure 5. Removing of the bead assembly

  8. Remove the protective cap covering the new bead.

  9. Mount the new bead assembly on the NPD lid. Be careful not to bump the bead on the sides of the lid or collector. 

  10. Replace the screws. Finger-tighten the first screw; tighten the remaining screws normally and then completely tighten the first screw. Do not overtighten the screws.

  11. Carefully bend the bead assembly cable 90°.

    KP665-AgilentNPD-InstallBead.png
    Figure 6. Installing of the bead assembly. 

  12. Reconnect the bead assembly cable to the NPD cable and twist the ring to lock the connection.


    Figure 7.  Reconnecting the bead assembly cable.

  13. Close the NPD cover and reinstall all GC covers. All covers must be closed to get a stable NPD baseline.

      
    Caution: Before you turn on the new NPD bead for the first time, manually set its voltage to a safe value so that the new bead is not destroyed. See the Turning on the NPD bead chapter in the Replacing the NPD Bead instruction guide for complete instructions.
  14. For the 7890A GC, configure the new bead.

    1. Set the bead type

    2. Review the Maximum Bead Voltage setting and adjust, if necessary.

    3. Review the Dry Bead and Auto Adjust Bead settings.

  15. Restore normal NPD operating gas flows.

  16.  With all gases on, heat the detector to 150 °C and hold for about 15 minutes, then increase the temperature to 250 °C and hold for 15 minutes. 

  17. Increase the temperature to operating value (310 to 320 °C recommended). Allow 15 minutes for equilibration.

  18. Check the NPD leakage current.

    1. Display the detector output.

    2. Verify that the output (leakage current) is stable and < 2.0 pA. If > 2.0 pA, verify bead installation and recheck. If still high, replace the ceramic insulators; inspect/replace the collector.

  19. Make sure that Adjust Offset is turned Off.

  20. After the temperature stabilizes at setpoint, set the initial Bead Voltage, depending on bead type:

    1. Blos bead: 0.0 to 0.5 V

    2. Ceramic bead (white or black): 0.0 to 2.0 V

  21.  After the bead reaches the initial voltage, begin to increase the voltage value in 0.05 V increments until the bead ignites. Wait about 10 seconds between each voltage adjustment. Monitor the detector output. 

    1. A new Blos bead typically ignites between 0.5 and 1.1 V.

    2. A new ceramic bead typically ignites between 2.8 to 3.2 V.

  22. When the bead ignites, the output will rise suddenly, then decrease towards a more stable value. It is best to allow the NPD to remain in this state without further adjustment for about 24 hours.

  23. Then you may adjust the bead voltage in small increments (0.05 to 0.1 V) until reaching the desired offset. With a clean environment, clean gas supplies, and low bleed column, a typical offset may decrease 6- 12 pA during a 24-hour period. 

  Learn how to effectively maintain your Agilent NPD:
GC-MULTI-2251r - NPD Maintenance e-learning course available from Agilent education