Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles Capillary Column Installation Quick Reference Guide
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Systems
  • Component :  GC Inlets GC System GC Detectors
  • Product Name :  6850 GC System 6890 GC System 7820 GC System 7890 GC System 8860 GC System 8890 GC System
  • Task :  Maintenance Operation
Related Articles
  • Recommended Parameters for GC Maintenance Methods

    This Information Applies To: Agilent GC systems Issue Before most maintenance procedures, the GC must be set to safe temperatures, pressures, or flows. You can create and save the maintenance method in the controlling software or the GC itself. A proper maintenance method will protect the user from injures, and the instrument from damage while performing maintenance. Tips : The Intuvo 9000, 8890 and 8860 GC systems will load the right instrument setpoints while performing guided...
  • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC

    This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet Issue Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet degrades over time due to accumulated contamination (e.g., baseline noise, ghost / artifact peaks, loss of reproducibility, carry over, loss of compound response or no compound response). Items Required The following Gold seals are available...
  • Agilent GC Detector Sampling (Data) Rate

    This Information Applies To: Agilent GC Systems Issue The data rate determines the number of sampling points across the chromatographic peak and the sample background. The data rate is based on the width of the analyte speak (in minutes) and the ratio of the analyte signal to the background noise. Background The data rate unit is hertz (Hz), where 1 Hz is one cycle per second. The data rate determines the number of data points taken across the width of the peak, and in the background...
  • Agilent Thermal Conductivity Detector Troubleshooting

    T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector) Issue Problems observed with thermal conductivity detector (TCD) results can have different underlying causes. Resolution Check that the TCD settings are within the recommended range, then assess the instrument for the following errors/problems: Hardware Fault Error No signal and/or the GC front panel screen displays Front/Back Detector Error. Cause: Filament is...
  • Bake Out the Agilent Thermal Conductivity Detector

    This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD) Issue Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces. Items Required Ferrule, no hole (p/n 5190-4054 ) Column nut, any of the following: Column nut, universal (p/n 5181-8830 ) Blanking plug, finger-tight style (p/n 5020-8294 ) Finger-tight column nut (p/n 5020-8292 ) Column nut, collared...

Capillary Column Installation Quick Reference Guide

Answer

This Information Applies To: All Agilent Gas Chromatographs, 6890, 6850, 7890, 7820, 8890, and 8860.


Issue:

A quick reference guide on the capillary column installation for the GC.

Inlets Detectors

Split/splitless inlet
Purged packed inlet
Multimode Inlet
Cool on Column (COC) Inlet
PTV Inlet
Volatiles Interface (VI) Inlet

FID / NPD
TCD - Standard Low Leakage Ferrules
TCD - Original Design Column Adapter
µECD
FPD

 


Steps to follow

 Warning 
Warning: The oven and/or inlet/detector may be hot enough to cause burns. Cool the GC down before column installation or if either is hot, wear heat-resistant gloves to protect your hands.

 

 Tip 
Tips: Wear clean, lint-free gloves to prevent contamination of parts with dirt and skin oils.

 

Inlets

Split/splitless inlet

  • Place a septum over the column, then the column nut and ferrule. Trim the end of the column with a column cutter.

  • Pull the column back so that 4-6 mm of the column is extending past the end of the ferrule (see Figure 1).

  • Thread the column nut and column into the inlet and tighten slightly past where the column grabs – retighten after heating. 


Figure 1
1. Ferrule, 2. Column nut, 3. Septum 


Purged packed inlet

  • Place a septum over the column, then the column nut and ferrule. Trim the end of the column with a column cutter.

  • Pull the column back so that 1-2 mm of the column is extending past the end of the ferrule (see Figure 2).

  • Thread the column nut and column into the inlet and tighten slightly past where the column grabs – retighten after heating.


Figure 2

 

Multimode Inlet

 Tip 
Tips: Make sure that the column adapter nut on the inlet base is fully threaded on and spinning freely – Collar Up! (see Figure 3)

Figure 3
  • Position the column so it extends 10-12 mm (for high boilers you can use 12-14 mm) above the end of the ferrule (see Figure 4). 

Figure 4
  • Tighten with two wrenches - 1/4” and 5/16” to prevent damage the inlet threads (see Figure 5).

Figure 5

Cool On Column (COC) Inlet

  • Insert the column all the way into the inlet until you feel the spring tension – do not withdraw.

  • Tighten with two wrenches - 1/4” and 5/16” to avoid damaging the inlet (see Figure 5).

PTV Inlet

  • There should be 17 mm of the column above the Graphpack ferrule – the Graphpack ferrule should be installed with the graphite end towards the inlet base (see Figure 6).

  • The column nut is slotted. Use a 5 mm wrench to tighten the fitting.


Figure 6  

 

Volatiles Interface (VI) Inlet

  • There is a longer column nut for the VI so that you don’t have to remove the inlet block. The column nut part number is p/n G3504-20504.

Figure 7

 

Detectors

FID/NPD 

  • Place a septum over the column, then the column nut and ferrule. Trim the end of the column with a column cutter.

  • Thread the column nut and column into the detector base. Gently insert the column into the detector until it bottoms, then withdraw 1 to 2 mm before tightening the nut.

     Tip 
    Tips: The dimensions shown of 48 and 68 mm are provided for narrow columns (<100 µm) which may slide all the way through the jet tip. (Figure 8)

Figure 8

 

TCD – Using the standard Low Leakage Ferrules

  • Place a 1/8” Swagelok nut and the back and front ferrules over the column. (Be sure to select the front ferrule size to match the column!) Trim the end of the column with a column cutter.

  • Thread the column nut and column into the detector base. Insert the column all the way into the detector jet until it stops, then withdraw 1 to 2 mm before tightening the nut. Retighten after heating.

Figure 9
1. Front ferrule, 2. Back ferrule, 3. Nut

TCD – Original Design Column Adapter

  • Install the column adapter to the 1/8-inch Swagelok TCD base using a graphite/vespel ferrule.

  • Thread the column nut and column into the adapter. Insert the column into the detector until it bottoms, then withdraw 1 to 2 mm before tightening the nut.



Figure 10
1. Ferrule, 1/8-inch, 2. Nut, 1/8-inch, 3. Capillary column adapter

 

µECD

  • Thread the column nut with column into the make-up gas adapter. Insert the column all the way into the mixing liner until it stops, then withdraw 1 to 2 mm - Use 2 wrenches to tighten the nut. Retighten after heating.

  • Be sure that the make-up gas adapter is installed with the 19 mm dimension shown (see Figure 11).


    Figure 11



    Figure 12  
     Tip 
    Tips: The 70 mm dimension shown is provided for narrow columns (<100 µm) which may go all the way through the mixing liner (see Figure 12).

FPD

  1. Install the column into the Column Measuring Tool provided.

  2. Trim the column such that 1 mm extends from the end of the tool.

  3. Tighten the column nut and mark the position of the column at the back of the nut.

  4. Install into the detector base.


Figure 13
1. Score column here, 2. Column measuring tool, 3. Ferrule, 4. Column nut, 5. The column length *

 

 Tip 
Tips: * The column length depends on the GC and FPD type. FPD+ for 7890B/8890/8860 is 137.5 mm, FPD for 7890A, and some 6890 is 145 mm, FPD for some early 6890 is 153 mm.

 

 Tip 
Learn how to effectively install GC capillary columns:
Installation Care and Maintenance of Capillary GC Columns e-learning available from Agilent education.
 