This Information Applies To: Agilent 6850, 6890, 7820, 7890, 8850, 8860, 8890, Intuvo 9000 GC Systems with a Thermal Conductivity Detector (TCD)
Issue
The TCD temperature, column (carrier), makeup, and reference flow rates need to be optimized to ensure proper operation and sensitive detection. This article guides the user on setting the correct temperature and gas flow rates on the TCD.
Steps to Follow
Detector temperature
|
Caution: The TCD temperature must not exceed the manufacturer recommended maximum column temperature limit as exceeding this limit will damage your column.
|
Gas flows
There are three gases flowing into the TCD (Figure 1):
Figure 1. TCD gas flows.
Left: Column effluent is forced away from the filament. TCD measures reference gas.
Right: Column effluent is forced toward the filament. TCD measures column flow sample peaks (if present).
1. Column flow 2. Make up gas flow 3. Reference gas switched towards the detector filament 4. Reference gas switched away from the detector filament.
Analytical column (carrier) gas flow. The typical column flow rate depends on the column diameter (Table 1):
|
Warning: Do not use hydrogen as the carrier gas in packed columns, and where possible, you can use it as the carrier gas in capillary columns.
Hydrogen is flammable and presents an explosion hazard when mixed with air in an enclosed space. The use of hydrogen as the makeup gas is not supported by the gas chromatograph hardware or software. |
Column type | Column size | Carrier flow rate, mL/min | ||
Hydrogen | Helium | Nitrogen | ||
Packed | 1/8-inch | Do not use | 30 | 20 |
1/4-inch | Do not use | 60 | 40 | |
Capillary | 0.05 mm id | 0.5 | 0.4 | NA |
0.10 mm id | 1.0 | 0.8 | NA | |
0.20 mm id | 2.0 | 1.6 | 0.25 | |
0.25 mm id | 2.5 | 2.0 | 0.5 | |
0.32 mm id | 3.2 | 2.6 | 0.75 | |
0.53 mm id | 5.3 | 4.2 | 1.5 |
Table 1. Recommended column flow rate
Multiply the total column + make up gas by ratio value to get the reference flow rate, e.g., 2.1*20 mL/min = 42 mL/min
Tips: The instrument itself maintains a minimal continuous reference flow of about 5 mL/min whenever the detector is enabled. Thus, any entered value less than about 5 mL/min has no effect on the actual flow rate.
|
See Table 2 for other worked examples.
Figure 2. Calculate the reference gas flow
1. Total column + makeup gas flow mL/min, 2. Ratio of reference gas flow for the total column + makeup gas flow
Column* | Column flow mL/min |
Makeup flow mL/min |
Combined mL/min |
Reference flow ratio (Figure 2) |
Reference flow mL/min |
Total detector flow mL/min |
Small bore capillary column: 25 m × 0.2 mm | 0.75 | 5 | 5.75 | 3 * combined flow | 3 × 5.75 = 17.25 | 5.75 + 17.25 = 23 |
Small bore capillary column: 25 m × 0.32 mm capillary | 10 | 2 | 12 | 2.25 * combined flow | 2.25 × 12 = 27 | 12 + 27 = 39 |
Large bore capillary column: 10 m × 0.53 mm capillary | 15 | 2 | 17 | 2.1 * combined flow | 2.1 × 17 = 35.7 | 17 + 35.7 = 52.7 |
Packed column: 1/8-inch stainless steel | 30 | 0 | 30 | 1.75 * combined flow | 1.75 × 30 = 52.5 | 30 + 52.5 = 82.5 |
Learn how to effectively operate and troubleshoot your Thermal Conductivity Detector (TCD):
GC-0GEN-1013s - GC Detectors Theory and Operation GC-7890-2281r - Agilent 7890 Series GC Thermal Conductivity Detector (TCD) Troubleshooting, e-learning courses available from Agilent Education |