This Information Applies To: Agilent OpenLab CDS
Issue
When analyzing a sample, the chromatogram may show baseline drift. The obtained signal may be caused by analytes as well as by dilution solvents, mobile phases, additives etc. To get a cleaner chromatogram with contribution of the analytes only, perform the blank subtraction.
Background
Blank signals are obtained from:
The new signal is calculated by subtracting the blank signal:
New signal = sample signal - blank signal
For extracted chromatograms using a specific wavelength:
New signal = sample signal at wavelength - blank signal at wavelength
Prerequisites
Steps to Follow
Figure 1. Choose which injections to perform blank subtraction
1. Processing Method window, 2. General, 3. Signals node, 4. Blank Subtraction tab, 5. Blank subtraction applied on
Figure 2. Blank column in Injection List
1. Last blank in the current injection, 2. Other blank option in the current injection under Blank column
Figure 4. Example of Blank Subtraction using a Last Blank in a Current Injection
Figure 5. Chromatogram with background noise removed
1. Before Blank Subtraction, 2. After Blank Subtraction
Learn how to effectively operate your Agilent OpenLab CDS:
HPLC-OLII-1131e - Agilent OpenLab CDS Data Analysis - Processing Methods e-learning course available from Agilent education |