This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software without Security Pack addon
Issue
The 2100 Expert software gives a counter mismatch error in the log book which may lead to data loss (blank lanes). Specifically, it occurs when the 2100 Expert software on the PC...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software
Issue
The DV 200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE) samples established by Illumina Inc. in 2016. It represents the percentage of RNA fragments larger than...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software, RIN NA, Error 4501, RNA 6000 Nano kit (p/n 5067–1511 ), RNA 6000 Pico kit (p/n 5067–1513 )
Issue:
The RIN is not displayed in the software or shows as NA. If the RIN is NA, the sample will be flagged in red and error 4501 will...
This Information Applies To: 2100 Bioanalyzer, 2100 Expert software, RNA 6000 Nano kit (p/n 5067-1511), RNA 6000 Pico kit (p/n 5067-1513)
Issue:
Which assay should be selected in the Agilent 2100 Expert software when running yeast total RNA samples on the Agilent 2100 Bioanalyzer system?
Resolution...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software, RNA assay, DNA assay, Protein assay
Issue:
The X-axis description toggle icon in the Agilent 2100 Expert software toolbar ( Figure 1 ) does not change the X-axis from time scale to base pairs, nucleotides, or kilodaltons...