This Information Applies To: Agilent SIPS 10/20
Issue
Operational tips for the Sample Introduction Pump System (SIPS) 10/20.
Background
SIPS is an integrated, efficient dilution system for flame AA and is suitable for use with most Agilent (flame) atomic absorption spectrometers. See Figure...
This Information Applies To: Agilent Flame AA Systems: 55B, 240FS, 280FS
Issue:
After cleaning and re-installing the flame atomic absorption system components, it is important to optimize the burner position, nebulizer uptake rate, flame conditions, glass impact bead , and mixing paddles . Proper...
This Information Applies To: Agilent atomic absorption systems (AA 240/280/55B/240Z/280Z)
Issue
Replacing hollow cathode lamps is a common maintenance procedure and can be used during troubleshooting to isolate the source of an error.
Steps to Follow
If you are developing a method using an...
This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to optimize the Agilent Flame AA for better instrument performance and response stability
Steps to follow:
Use Figure 1 as reference for the procedure on the Flame AA optimization...
This Information Applies To: Agilent GTA 120 and Zeeman Atomic Absorption instruments
Issue
How to replace and align graphite tubes. Replacement of graphite tubes are a routine maintenance item for all Graphite Furnace instruments.
Items Required
Graphite tubes suitable for your application...