Hi all,
I have problem in GC6890 ,I can not connect with open lab and a message appears licensing error
any help.
Thanks
Dheya
Hi all,
I have problem in GC6890 ,I can not connect with open lab and a message appears licensing error
any help.
Thanks
Dheya
Hello,
I added tags to this post to increase visibility.
OpenLab licenses are tied to two things on your PC: the computer name and the MAC address. If this license has worked in the past but now has stopped working, then the first things you want to check are if the computer name has changed or if the MAC address that was used in the license file is no longer detected in Windows (maybe a NIC was replaced).
Details on how you can check this information can be found in this FAQ document: FAQ: How to Generate the License File in SubscribeNet | Agilent
If there was a change in the computer name and/or MAC address, you will need to return your current license and generate a new one with the PC's current computer name and MAC address. The following FAQ outlines how to return your license in subscribenet: How do I Return an Existing OpenLAB CDS License in SubscribeNet? | Agilent
Hopefully this helps.
Hello,
I added tags to this post to increase visibility.
OpenLab licenses are tied to two things on your PC: the computer name and the MAC address. If this license has worked in the past but now has stopped working, then the first things you want to check are if the computer name has changed or if the MAC address that was used in the license file is no longer detected in Windows (maybe a NIC was replaced).
Details on how you can check this information can be found in this FAQ document: FAQ: How to Generate the License File in SubscribeNet | Agilent
If there was a change in the computer name and/or MAC address, you will need to return your current license and generate a new one with the PC's current computer name and MAC address. The following FAQ outlines how to return your license in subscribenet: How do I Return an Existing OpenLAB CDS License in SubscribeNet? | Agilent
Hopefully this helps.
Thank you for your response
Today when you open the GC we noted the increase returned valid but not connected and the Remote LED off
we tried to configuration instrument but a message appears(failed to connect to 6890 at 100.100.100.101)but the GC IP:100.100.100.101,when Checked configuration not configuration
YOUR HELP.
Hello,
Are you able to ping the GC at 100.100.100.101? If not, then you might want to verify the IP address assignment on the instrument and that the PC and GC are on the same subnet.