Clear calibration during process doesn't delete old data

Hi, 

I try to understand why I have old injections data for only some of my target calibrated compounds and other are ok. I have no idea. 

First here how I usually proceed to my calibration with bracketing mode at "none".

When I check the curve, the majority of my 30 calibrated compounds have one calibration point like it is supposed to be, but some have old or more recent injection data when we look at those curve points.

How can this happen? How can this only happen to some of my compounds. 

The majority of them are ok with one point only. 

THank you for your help. 

 

Parents Reply Children
  • HI Martin, I took some injection to the initial Result set and I click on the "add selected injections to new result set" as I did often in the past. 

    Usually it went well. 

    I will try what you have asked me to do and come back with the result. 

    thank you again for your help.

  • OK here is my response. 

    I confirm that for this particular folder, it didnt have the acquisition method for an unknown reason. I have checked other folders date and the acquisition method is there. 

    I am confused because I did the process and result id the same way. 

    SO I did transfer the process method in the problematic folder and I had the Reportableinformation. I reprocess to see if the problem is fixed but it was the same. 

    It must be something I did different for this folder but I don't know why. 

    Have you ever seen that ? 

  • Hello,

    When I tested on my system all the acquisition methods are copied over from the original result sets. Likely on your system there was some issue when creating the new result set in this case. You will not be able to copy the method into the folder, as I would imagine the manifest file is not pointing at the method in question. Are you combining results sets that have the same acquisition method? 

    Marty

Was this helpful?