Hello I am using 1290 iQ's with open access (openlab CDS version 2.6), sequence is pausing on missing vials which he is not supposed to do I checked the configuration of vial sampler and its okay ( there is no tick on checkbox).
Hello I am using 1290 iQ's with open access (openlab CDS version 2.6), sequence is pausing on missing vials which he is not supposed to do I checked the configuration of vial sampler and its okay ( there is no tick on checkbox).
Hi furqan.khan
The tick box needs to be ticked to ignore missing vials. Sounds like your 1290 is doing what it should do,
/Andy
Hi Andy, thanks for your reply sorry I was not very clear but it shouldn't be ticked. I am attaching the screenshot of configuration just to make more clear. Thanks.
Hi furqan.khan
Ah, so that's in Masshunter Walkup. I was actually referring to the option directly in the OpenLab software. If you right click on the Sampler in OpenLab Instrument Status and select the top menu item, "Control", you'll see another option to ignore missing vials.
Its possible that you might need to make sure that the options in Walkup and OpenLab.
Its really confusing that in Walkup the option is to "Stop on missing vial" and in Openlab its the other way around "Ignore missing vial".
Hope that is a little clearer.
/Andy
Hello furqan.khan ,
Yes, as Andy Wright mentions the option to continue on missing vial needs to be enabled in both WalkUp and the host data system. If you have that option set correctly in both WalkUp and the host data system and it still pauses or stops, then you would need to check the logs to see if it is being caused by OL not being ready or by WalkUp and then troubleshoot further from there.
Hello furqan.khan ,
Yes, as Andy Wright mentions the option to continue on missing vial needs to be enabled in both WalkUp and the host data system. If you have that option set correctly in both WalkUp and the host data system and it still pauses or stops, then you would need to check the logs to see if it is being caused by OL not being ready or by WalkUp and then troubleshoot further from there.
Hello howard,
Thanks for your reply I already mentioned to Andy Wright what I did but its not happening at the moment. we have another 1290 with Mass hunter Walkup but without mass spectrometer and its working fine he is skipping the vial without pausing the sequence. It could be installation issue or something with Mass Spec but I am not too sure. Anyway thanks for your time and feedback, I appreciate it.