I tried to copy EIC Signal description items near each other and I don't know how to avoid this problem showed below.
As you can see, the three chromatogram is the same compound. How can I avoid this repetition?
thank you for your help.
I tried to copy EIC Signal description items near each other and I don't know how to avoid this problem showed below.
As you can see, the three chromatogram is the same compound. How can I avoid this repetition?
thank you for your help.
Hello,
Try the MS Chromatogram Flowlayout object instead of the individual chromatograms.
Marty
Hi Martin. I have the same problem. I don't know how to manage the group repeating so we can have different compound name on chromatogram lighted in gray.
Have you try it? It seems that I don't have the privilege to do so.
I'm not sure what you mean. Are you unable to edit the report?
Yes I can edit report buy I can't do any changes with those statistical results. So that's why I asked you if you have try to change the order.
No, you don't need to change the folder or anything like that. Just select the column in the table on the report template, right click, click on "Column properties", go to the aggregates tab, then follow my instructions above. If you're still struggling, I can provide some snapshots in a few minutes.
Oh I see. It works now. Thank you.
There are two kinds of multipliers, one for compound in the cal table, and one (ok it's 5 :) ) in the injection list for the samples. The compound multiplier can not be displayed with a single variable, but there is a trick :)
At the last standard injection use this in a textbox or a table value: = Compound_Concentration/Compound_Amount*iif(Sample_Amount<>0,Sample_Amount/100,1)
For every other std injection, except the last one, it will give a false MP
There are two kinds of multipliers, one for compound in the cal table, and one (ok it's 5 :) ) in the injection list for the samples. The compound multiplier can not be displayed with a single variable, but there is a trick :)
At the last standard injection use this in a textbox or a table value: = Compound_Concentration/Compound_Amount*iif(Sample_Amount<>0,Sample_Amount/100,1)
For every other std injection, except the last one, it will give a false MP