I tried to copy EIC Signal description items near each other and I don't know how to avoid this problem showed below.
As you can see, the three chromatogram is the same compound. How can I avoid this repetition?
thank you for your help.
I tried to copy EIC Signal description items near each other and I don't know how to avoid this problem showed below.
As you can see, the three chromatogram is the same compound. How can I avoid this repetition?
thank you for your help.
Hello,
Try the MS Chromatogram Flowlayout object instead of the individual chromatograms.
Marty
Hi Martin. I have the same problem. I don't know how to manage the group repeating so we can have different compound name on chromatogram lighted in gray.
Hello,
The compound multiplier field in IR was added in CDS 2.7 to allow reporting of that value from the method per compound. You can back calculate that multiplier value from the compound amount and compound concentration but will need to take into account any other sample dilution factors, sample multipliers, and sample amounts used in the concentration calculation. Below is a custom calculation that will back calculate the compound multiplier for %mass calibrations where the dilution factors are used as multipliers. You really need 4 calculations to cover all calibration and dilution factor types. You should not enter 0 for the sample amount when using %mass calibrations as your concentration values will not be calculated.
Marty
Compound_Concentration/Compound_Amount/CurrentSample.Multiplier(1)/CurrentSample.Multiplier(2)/CurrentSample.Multiplier(3)/CurrentSample.Multiplier(4)/CurrentSample.Multiplier(5)/CurrentSample.DilutionFactor(1)/CurrentSample.DilutionFactor(2)/CurrentSample.DilutionFactor(3)/CurrentSample.DilutionFactor(4)/CurrentSample.DilutionFactor(5)/100*Sample_Amount
Hello,
If you CDS 2.7 the compound_mulitplier is present as an IR field. This screenshot below is from CDS 2.7 it shows the field in IR and CC.
Marty
Hello,
One option would be to stop using mass% and simply do the compound_cocentration/sample_amount calculation in IR or custom calculator.
Marty Adams
I will still need to show a mulitplier that will be needed to change my amount in ug in concentration expressed as PPM (ug/g).
no?
I am not using any dilution factor for now. Only one mulitplier for this calculation who is doing well with the % mass since the formula is the same for all compound.
Hello,
One other question, it looks like you are using 10 as the multiplier for all the compounds in your table. Why not just place 10 in one of the sample multipliers 1-5 for all the injections and set the compound multipliers back to 1. Typically, the compound multiplier is used when you need to apply a different factor to each compound for something like standard purity. In your version, you can get the sample multiplier as I detailed previously.
Marty
you're right. I will use the multiplier 1 to avoid this 2.6 version problem. Very good idea.
you're right. I will use the multiplier 1 to avoid this 2.6 version problem. Very good idea.