EIC report with smaller chromatogram to reduce the number of pages in the report

I tried to copy EIC Signal description items near each other and I don't know how to avoid this problem showed below. 

As you can see, the three chromatogram is the same compound. How can I avoid this repetition? 

thank you for your help. 

Parents Reply
  • Hello,

    The compound multiplier field in IR was added in CDS 2.7 to allow reporting of that value from the method per compound. You can back calculate that multiplier value from the compound amount and compound concentration but will need to take into account any other sample dilution factors, sample multipliers, and sample amounts used in the concentration calculation. Below is a custom calculation that will back calculate the compound multiplier for %mass calibrations where the dilution factors are used as multipliers. You really need 4 calculations to cover all calibration and dilution factor types. You should not enter 0 for the sample amount when using %mass calibrations as your concentration values will not be calculated.  

    Marty

    Compound_Concentration/Compound_Amount/CurrentSample.Multiplier(1)/CurrentSample.Multiplier(2)/CurrentSample.Multiplier(3)/CurrentSample.Multiplier(4)/CurrentSample.Multiplier(5)/CurrentSample.DilutionFactor(1)/CurrentSample.DilutionFactor(2)/CurrentSample.DilutionFactor(3)/CurrentSample.DilutionFactor(4)/CurrentSample.DilutionFactor(5)/100*Sample_Amount

Children
No Data