Area Count Limits

Hello. I'm running OL Chemstation C.01.10. Can you set area count limits in the software, whether by percentage or a set range? Ultimately, I'd like to flag an internal standard area count on a report if it falls outside of the set limits, if possible.

Ed

Parents
  • Edward,

    You would need to do this as a custom calculation in intelligent reporter. Is this for a single injection or a summary of multiple injections? The example below, I did for a customer project where the report averaged all the IS injections, calculated limits based on a percent of that average area, and then applied those limits using pass/fail logic and conditional formatting. 

    Marty Adams

Reply
  • Edward,

    You would need to do this as a custom calculation in intelligent reporter. Is this for a single injection or a summary of multiple injections? The example below, I did for a customer project where the report averaged all the IS injections, calculated limits based on a percent of that average area, and then applied those limits using pass/fail logic and conditional formatting. 

    Marty Adams