Openlab 2.4 Installation Error

Hi

 

I'm having an important issue with Openlab CDS 2.4 File based Workstation.

 

Installing v2.3 goes without any error, and I can completely connect and control a GC 6890N.

 

 

When updating to 2.4, I get a Install error on the GC/MS line

 

Running the Software Verification tool give me no error. Every thing says "OK" (logs will be included in this thread)

 

And then, on version 2.4, I can connect to my GC 6890N, see the online signal, import a method from the GC, but I'm totally UNABLE TO SEND A METHOD to the GC. 

This update to v2.4 make my GC unusable.

 

I tried to repair the installation with no success. Deleting v2.3 and installing 2.4 without going from 2.3 gives the same error. I also tried to install on a different HDD drive, but still same error.

 

I don't know what to do. Why is 2.3 ok and 2.4 fails ?

 

I'm including the GC/MS error log if you can see what goes wrong.

 

Thanks for your help.

 

-Francois

attachments.zip
Was this helpful?