Openlab 2.4 Installation Error

Hi

 

I'm having an important issue with Openlab CDS 2.4 File based Workstation.

 

Installing v2.3 goes without any error, and I can completely connect and control a GC 6890N.

 

 

When updating to 2.4, I get a Install error on the GC/MS line

 

Running the Software Verification tool give me no error. Every thing says "OK" (logs will be included in this thread)

 

And then, on version 2.4, I can connect to my GC 6890N, see the online signal, import a method from the GC, but I'm totally UNABLE TO SEND A METHOD to the GC. 

This update to v2.4 make my GC unusable.

 

I tried to repair the installation with no success. Deleting v2.3 and installing 2.4 without going from 2.3 gives the same error. I also tried to install on a different HDD drive, but still same error.

 

I don't know what to do. Why is 2.3 ok and 2.4 fails ?

 

I'm including the GC/MS error log if you can see what goes wrong.

 

Thanks for your help.

 

-Francois

attachments.zip
Parents
  • Hi there,

     

    revision 2.4 has several issues with their drivers, here are the workarounds:

     

    1) GCMS Driver installation failed

    Problem:  GCMS driver has problem with trusted certificate on offline machines

    Workaround: Edit Windows Registry -> type Start -> Run -> regedit

    Expand path:

    [HKEY_LOCAL_MACHINE\SOFTWARE\Policies\Microsoft\SystemCertificates\AuthRoot]

    and change value for DisableRootAuto... from 1 to 0.

    Do not restart PC and try installation again. It should go through GCMS driver installation step.

     

     

    2) 6890x instruments does not react for Downloading method. Cannot Acquire analyses

    Problem: New GC Driver 3.0xx and 3.0xx SR1 has issue on Windows 2012 and probably Windows 10. This is a bug which should be fixed by Agilent soon.

    Workaround: Download and install GC Driver B.01.04 HF1 from OpenLAB CDS 2.3 or from Subscribenet. This driver works correctly.

  • M. majkee, you are a savior  

     

    Thank you so much for your input.

    On a French Installation of Windows, using English Canada as the Region and language, I am indeed able to control our GC 6890N on Openlab 2.4 with the B.01.04 HF1 Drivers. 

     

    I'm having issues with automatic printing and export of report now, but that's so much less of an issue for me now !

     

    Thanks again so much.

     

    -Francois.

Reply
  • M. majkee, you are a savior  

     

    Thank you so much for your input.

    On a French Installation of Windows, using English Canada as the Region and language, I am indeed able to control our GC 6890N on Openlab 2.4 with the B.01.04 HF1 Drivers. 

     

    I'm having issues with automatic printing and export of report now, but that's so much less of an issue for me now !

     

    Thanks again so much.

     

    -Francois.

Children
No Data
Was this helpful?