Hello team,

 

I wanted to consult the following: an HPLC 1260, I had to reinstall the software. Once I did it, after creating a new method and giving the order to inject, the software turns blue and immediately turns green and does not inject...

 

At the bottom there is a message that says: System.Access.Violation.Exception Attempt to read or write to the protected memory. This often indicates that there is another damaged memory.

 

What can I do?

Parents Reply
  • Hello, my friend,

     

    After checking the logbook on the ChemStation, note that the method did not run due to an error in the G4212B detector indicating: Start Failed (temperature HW).

     

    After that, I checked the instrument with the Lab Advisor, and the Logbook showed the error: Heater at fan assembly failed (EE 1073, 0).

     

    Do you have any idea which part it could be? I suspect the fan cooler has a resistor. 