Auto run macros failed in Chemstation 1.07 software .Please help how to resolved it. It's urgent ....
Auto run macros failed in Chemstation 1.07 software .Please help how to resolved it. It's urgent ....
Hi Ryoboyle,
Here I have attached the instrument log. But one thing I forgot to informed you that yesterday I have already delete old instrument and create new one. So previous instrument log also deleted. But this problem is not solve yet.
Below this log file link
Thank you! Unfortunately, there was not really any additional information in that log file since it is a new instrument. At this point there are really only three things I can think of to try:
1) You can run the Software Verification Tool (found in Windows by going to Start>All Programs>Agilent Technologies>Software Verification Tool). This will let you know whether there are any missing or potentially corrupted files in your OLCDS ChemStation installation.
2) You can run a repair from disk 1. Make sure you are an administrator in Windows and then right-click on the setup.bat file on disk and select to 'Run as an Administrator'. This will launch the Master Installer where you can find the repair option in the Maintenance tab.
3) If all else fails, uninstall and reinstall the software.
Hi choushik, do you still need help on this issue? If so, please come back and let us know so we can continue trying to assist. If the problem was resolved by one of the answers provided, please mark it as "Correct Answer."
Reinstalled this software and solved this problem. Thanks