Find maintenance, best practice, and troubleshooting articles.
Hello jwhiddon,
please look for the autotune log and rename it. The location of the log depends on the software you're using. In MassHunter this should be C:\MassHunter\GCMS\1,2,3,4\Logs\autotunelog.ei.csv.
In MSD Chemstation it should be C:\msdchem\1\5975\eitunes\atune.csv.
If you rename or move this file, the software will create a new one. The tuneplot will start with new data and the error message is gone.
Best,
Michael
Thank you Michael! I knew the file must have been full, but I wasn't sure if the software would automatically create a new one if I moved the old file.
Thank you Michael! I knew the file must have been full, but I wasn't sure if the software would automatically create a new one if I moved the old file.