Alguém sabe como resolver esse status de Energia e Fonte em vermelho?
O Cary 630 não permite nenhuma análise com este status!
Obrigado
Alguém sabe como resolver esse status de Energia e Fonte em vermelho?
O Cary 630 não permite nenhuma análise com este status!
Obrigado
Hi JORGE.AMARAL
Thank you for your question. I’ve used an online tool to translate your post. English is the official language of the Agilent Community, and the platform does not support translating a post from local language to English. You can view pages in your local language by clicking the “globe” icon in the upper-right corner and selecting your language. Learn more Agilent Community Translation - Community Help - Getting Started - Agilent Community. See translation of the original post below.
Does anyone know how to resolve this Energy and Source status in red?
The Cary 630 does not allow any analysis with this status!
Cheers,
Kristen
Hi JORGE.AMARAL
It seems that the source should be replaced.
Let me know if it everything is ok after replacing it.
Regards
Hi guys!
Sorry for sending the message in my native language!
I ask the staff to replace the source.
thank a lot!
The User manual has both the instructions to replace and the part number for the source. We also recommend changing the desiccant too. The User manual is available online at www.agilent.com From the drop-down menu, located left of the search window, choose User manual, and type 630 in the search window.
Hello, Jorge! Have you ever try to change gain? Go to Gain Adjust e try to put close to 25000. If it still continue red, try to purge your IR with nitrogen gas. If you prefer to speak in pt-br send me a mensager.