We have a Cary 60. We are doing preventive maintenance, but the noise test fails and it took up to three or four times to get a successful result.
What can we do to improve this and get a passed result the first time?
Hi there!
We have a Cary 60. We are doing preventive maintenance, but the noise test fails and it took up to three or four times to get a successful result.
What can we do to improve this and get a passed result the first time?
Was it run with micro-cuvettes instead of macro-cuvettes?
I learned recently that the closeness of some micro-cuvette walls might refract (for lack of a better word) additional light.
And of course the quality must be QS aka QG.
Was it run with micro-cuvettes instead of macro-cuvettes?
I learned recently that the closeness of some micro-cuvette walls might refract (for lack of a better word) additional light.
And of course the quality must be QS aka QG.